If we are getting split peak in Chromatogram,what we can do?

Answers were Sorted based on User's Feedback



If we are getting split peak in Chromatogram,what we can do?..

Answer / alpesh patel

It means yr column is contaminated & the column needs
proper washing with suitable solvents. You can also
regenarate the column by proper washing with
polar solvent -> semi polar solvent -> non polar solvent ->
semi polar solvent -> polar solvent.

Is This Answer Correct ?    18 Yes 4 No

If we are getting split peak in Chromatogram,what we can do?..

Answer / mr. bujji reddy kanchi

The peak split involves both mobile phase and stationery
phase. The pH of the mobile phase accuracy depends.

Mainly Stationery phase deterioration. Column has to be
regenerated according to the Column certificate
recommondation solvents.

Is This Answer Correct ?    6 Yes 0 No

If we are getting split peak in Chromatogram,what we can do?..

Answer / satyasri

if u observe peak splitting,first u go to change(reduce) injection volume(means volume over load)...dn u go for buffer srength(means conc)...dn u go for pH of the buffer...dn only u go for checking of column performance(means whether u hv to check column was washed properly from previos one)........k

Is This Answer Correct ?    2 Yes 1 No

If we are getting split peak in Chromatogram,what we can do?..

Answer / jayalakshmi

If the sample diluent is not compatable with mobile phase,
it could be the reason for peak splitting. Using mobile
phase as sample diluent would resolve the issue.

Is This Answer Correct ?    1 Yes 0 No

If we are getting split peak in Chromatogram,what we can do?..

Answer / g.balaji

Main cause : That means u r column is contaminated
absorbant ion of impurites in the column.

Solution :
1)Wash calumn:-By polar solvent -> semi polar solvent -> non
polar solvent ->semi polar solvent -> polar solvent.
2)Check column performance
3)Replace column

Is This Answer Correct ?    0 Yes 3 No

If we are getting split peak in Chromatogram,what we can do?..

Answer / umesh patel

BECOZ OF COLUMN OVERLOADING
DECREASE THE INJECTION VOLUME AND CHECK OUT

ALSO
IT HAPPENS THAT IMPURITIES FROM PREVIOUS INJECTIONS ELUTED

RUN BLANK WITH PROPER SOLVENT AND THEN WITH MOBILE PHASE

Is This Answer Correct ?    1 Yes 5 No

If we are getting split peak in Chromatogram,what we can do?..

Answer / n.s.r

In Method development time or already validated method first we conform that already validated method it's ok but in development time we go on peak purity and finalise column problem or if any impurities is present in main peak.Thank you ANY WAY YOU WILL answer that question.

Is This Answer Correct ?    0 Yes 6 No

Post New Answer

More Analytical Chemistry Interview Questions

what are the sources in ELSD,PDA, RI and UV Detectors?

1 Answers  


which effect is mainly responsible for chromatographic separation when silica is the statinary phase at TLC? thanks in advance

1 Answers  


which chromatographic technique is used for softening hard water?

1 Answers   Laurus,


in determination of(acetone) moister content we are using methanol replaced by pyredine.what hapend in that reaction.

2 Answers   Glenmark, Neuland,


Loss of linearity in tkn analysis, possible reasons?

0 Answers  






Why will a solution of aspirin without acid or basic hydrolysis give yellow colour with a solution of Ferric chloride?

1 Answers  


What is the calibration procedure for Malvern MS2000 particle size analyzer and acceptance criteria?

2 Answers   Reddy Labs,


What is the optimum period of storage of indicators?

1 Answers   Orchid Chemicals and Pharmaceuticals,


if peak get problamatic then what require to do?

0 Answers  


why we are using k2cr207 in uv calibration ? why we r not using kmno4?

1 Answers   Glochem, Sidmak Laboratories,


Why we have to take 42gm of sodium hydroxide instead of 40gm for preparation of 1M Sodium hydroxide as per Pharmacopia? Whether it means the formula for calculating molarity was wrong? 

5 Answers   Claris Otsuka, Ranbaxy,


In gas chromatographic analysis sometimes we use headspace, why it is required instead of manually injection?

1 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)