Answer / alpesh patel
It means yr column is contaminated & the column needs
proper washing with suitable solvents. You can also
regenarate the column by proper washing with
polar solvent -> semi polar solvent -> non polar solvent ->
semi polar solvent -> polar solvent.
| Is This Answer Correct ? | 18 Yes | 4 No |
Answer / mr. bujji reddy kanchi
The peak split involves both mobile phase and stationery
phase. The pH of the mobile phase accuracy depends.
Mainly Stationery phase deterioration. Column has to be
regenerated according to the Column certificate
recommondation solvents.
| Is This Answer Correct ? | 6 Yes | 0 No |
Answer / satyasri
if u observe peak splitting,first u go to change(reduce) injection volume(means volume over load)...dn u go for buffer srength(means conc)...dn u go for pH of the buffer...dn only u go for checking of column performance(means whether u hv to check column was washed properly from previos one)........k
| Is This Answer Correct ? | 2 Yes | 1 No |
If the sample diluent is not compatable with mobile phase,
it could be the reason for peak splitting. Using mobile
phase as sample diluent would resolve the issue.
| Is This Answer Correct ? | 1 Yes | 0 No |
Main cause : That means u r column is contaminated
absorbant ion of impurites in the column.
Solution :
1)Wash calumn:-By polar solvent -> semi polar solvent -> non
polar solvent ->semi polar solvent -> polar solvent.
2)Check column performance
3)Replace column
| Is This Answer Correct ? | 0 Yes | 3 No |
Answer / umesh patel
BECOZ OF COLUMN OVERLOADING
DECREASE THE INJECTION VOLUME AND CHECK OUT
ALSO
IT HAPPENS THAT IMPURITIES FROM PREVIOUS INJECTIONS ELUTED
RUN BLANK WITH PROPER SOLVENT AND THEN WITH MOBILE PHASE
| Is This Answer Correct ? | 1 Yes | 5 No |
Answer / n.s.r
In Method development time or already validated method first we conform that already validated method it's ok but in development time we go on peak purity and finalise column problem or if any impurities is present in main peak.Thank you ANY WAY YOU WILL answer that question.
| Is This Answer Correct ? | 0 Yes | 6 No |
what is full form of RVT
How to calculate mass balance in RS after degradation
what is difference between zero air gas and o2 gas
1 Answers IPCA, Vasudha Pharma,
what is karl fischer principle and reaction ,describe it
5 Answers Jubilant Organosys, Microsoft,
which one is better separation in high pressure liquide chromatogram orlowpressure chromato gram in hplc
How do we quantify crystaline and amarpous forms by using (NMR, XRD)spectroscopic techniques? Which any others instruments are useful for this quantification? explain
0 Answers Graviti Pharmaceuticals, Vimta Labs,
WHY POTASSIUM HYDROGEN PTHALATE IS USED FOR PERCHLORIC ACID? What is difference btw UV detector & RID detector? what is dissolution? What is unit of KF? what is the significant of pH in HPlc?
what is relationship between absorption and transmittance in u.v.?
What are the different zones in stability according to ICH?
Why we show ammonia vapours to tlc plate?to which compounds?
What is the name of the detector used in TOC?
Tell me something about Vitamin A test method by HPLC