Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...


If we are getting split peak in Chromatogram,what we can do?

Answers were Sorted based on User's Feedback



If we are getting split peak in Chromatogram,what we can do?..

Answer / alpesh patel

It means yr column is contaminated & the column needs
proper washing with suitable solvents. You can also
regenarate the column by proper washing with
polar solvent -> semi polar solvent -> non polar solvent ->
semi polar solvent -> polar solvent.

Is This Answer Correct ?    18 Yes 4 No

If we are getting split peak in Chromatogram,what we can do?..

Answer / mr. bujji reddy kanchi

The peak split involves both mobile phase and stationery
phase. The pH of the mobile phase accuracy depends.

Mainly Stationery phase deterioration. Column has to be
regenerated according to the Column certificate
recommondation solvents.

Is This Answer Correct ?    6 Yes 0 No

If we are getting split peak in Chromatogram,what we can do?..

Answer / satyasri

if u observe peak splitting,first u go to change(reduce) injection volume(means volume over load)...dn u go for buffer srength(means conc)...dn u go for pH of the buffer...dn only u go for checking of column performance(means whether u hv to check column was washed properly from previos one)........k

Is This Answer Correct ?    2 Yes 1 No

If we are getting split peak in Chromatogram,what we can do?..

Answer / jayalakshmi

If the sample diluent is not compatable with mobile phase,
it could be the reason for peak splitting. Using mobile
phase as sample diluent would resolve the issue.

Is This Answer Correct ?    1 Yes 0 No

If we are getting split peak in Chromatogram,what we can do?..

Answer / g.balaji

Main cause : That means u r column is contaminated
absorbant ion of impurites in the column.

Solution :
1)Wash calumn:-By polar solvent -> semi polar solvent -> non
polar solvent ->semi polar solvent -> polar solvent.
2)Check column performance
3)Replace column

Is This Answer Correct ?    0 Yes 3 No

If we are getting split peak in Chromatogram,what we can do?..

Answer / umesh patel

BECOZ OF COLUMN OVERLOADING
DECREASE THE INJECTION VOLUME AND CHECK OUT

ALSO
IT HAPPENS THAT IMPURITIES FROM PREVIOUS INJECTIONS ELUTED

RUN BLANK WITH PROPER SOLVENT AND THEN WITH MOBILE PHASE

Is This Answer Correct ?    1 Yes 5 No

If we are getting split peak in Chromatogram,what we can do?..

Answer / n.s.r

In Method development time or already validated method first we conform that already validated method it's ok but in development time we go on peak purity and finalise column problem or if any impurities is present in main peak.Thank you ANY WAY YOU WILL answer that question.

Is This Answer Correct ?    0 Yes 6 No

Post New Answer

More Analytical Chemistry Interview Questions

how can we calculate void volume?give its importance in brief

1 Answers   Natco,


what is principal of karl ficher instrument!

12 Answers   Alkem Labs, USV,


what is mean by covalidation

0 Answers  


What is definition of RRT and what is the difference between RT and RRT in HPLC

2 Answers  


what is the basis of separation in chromatography?

1 Answers   Anthem,


what is the difference between LOSS ON DRYING IN ORDINARY OVEN TECHNIQUE AND VACUUM OVEN TECHNIQUE

6 Answers  


define polar and non polar solvent

2 Answers  


How require to perform linearity as per ANVISA? What are the acceptance criteria?

0 Answers  


In HPLC Chromatographic separation Resolution equation is Rs = 1.18 (tR2 –tR1)/wh1+Wh2, Where, tR2 and tR1 is run time of eluent ratios of the distances from the point of application to the centers of the spots and the distance travelled by the solvent front from the point of application (retardation factor), and Wh1and Wh2 , is peak width of at half height. What is 1.18 ?

1 Answers  


What is discriminating media? How can u justify/select that which media (pH etc...)is suitable for your dissolution study of your compound?

5 Answers   Natco, SUN,


what is difference between Performance check and Calibration?

1 Answers  


Why Urasil and Caffeine use for HPLC wavelength accuracy calibration. which is most suitable, Why? other than these which can be used?

7 Answers   Cipla,


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)