Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...


As per OOT for any product what is calculation and how to
invetigate the OOT

Answers were Sorted based on User's Feedback



As per OOT for any product what is calculation and how to invetigate the OOT..

Answer / s jagadeesh chandra

This is only a briefe answer.

OOT means out of trend.

Definition:

In case of routene analysis any result of quantitative
chemical test that do not fall with in the defined band in
comparison to a specification.

Investigation:

Investigation procedure is same as that of OOS(Out of
specification).

OOt result in qc laboratory
.
Report with in 24 hours to qa&supervisor.
.
phase I laboratory investigation is carried out.
.
If any laboratory error is identified then it is termed as
assiagnable cause(ie system problem identified,sample
preparation,dilution in correct,improper transfer of
solvent,analyst etc etc...).There are many check points are
there to investigate.The investigator will go through the
check points.
.
After assiagnable cause identification corrective and
preventive action will be taken,and order for re testing
the original sample or another on same or another equipment
(if system performance is suspected).The person other than
who obtain the initial results should re test the
sample.After re testing results are to be evaluated and
conclude the investigation.

Re testing plan:2 sample preparetions to be done and analyse

Different different OOT limits are there for different
different tests.

For more details mail me.

Is This Answer Correct ?    16 Yes 0 No

As per OOT for any product what is calculation and how to invetigate the OOT..

Answer / manojmalik

Investigation of OOT ,like same as per the OOS
investigation ,Obseved the OOt as per the regrassion
cofficient .based ob 1,2and 3 sigms value .

Is This Answer Correct ?    3 Yes 1 No

As per OOT for any product what is calculation and how to invetigate the OOT..

Answer / vamsidhar reddy. b

1. Tabulate the data, calculate the average and standard deviation.
• Use excel spread sheet for average & Standard deviation calculation.
• Standard deviation will be multiplied by 3 to get the 3 sigma (3 )) value.
• Calculate ± 3 Sigma
• Maximum & Minimum limits will be fixed by the formula Average ±3 Sigma
• If the obtained maximum value is above the spec. Limit, specification limit shall be taken into consideration.
• The obtained minimum value may be below the lower spec. limit.
• The maximum and minimum limits as explained above shall be taken as the trend range for upper and lower limits.
• Any obtained value is found to be out of this range will be considered as Out of Trend (OOT) value.
2. Based on preliminary investigation OOT Results should be reconfirmed with same analyst in duplicate.
3. If the result of repeated analysis is in line with the result of 1st analysis, then results of 1st analysis shall be reported and 2nd analysis is considered ‘For confirmation only’ and further investigation is carried out for the cause of OOT result.
4. If the result of repeated analysis is line with previous trend, then result of 2nd analysis shall be reported and no further investigation is carried out with respect to the product.

Is This Answer Correct ?    1 Yes 0 No

Post New Answer

More Analytical Chemistry Interview Questions

why we can use only 4 and 7 and 9 buffers in pH calibration

1 Answers   Chola, Cipla,


why kf factor coming 0.5

5 Answers   Alkem Labs,


What is rs test why we are performing rs test?

0 Answers  


Qc interview questions

0 Answers   Apotex,


If we do accuracy at same concentration at which linearity planned,what is the need to do linearity separately?

0 Answers  


What is a base line?

2 Answers  


What is the Formula for coreletion coefficient in plhplc calibration

0 Answers  


What is related substance by HPLC impurity limits as per USP?

0 Answers  


how to calibrate hplc & gc

0 Answers   Cipla,


How can we calculate "confidence interval" in analytical method validation? Pl. explain with example.

0 Answers   Cipla, Rampex Labs,


what is the difference between capillary and glass column in GC

1 Answers  


What is difference between silica used in TLC and HPLC column

4 Answers   Cipla, Meghmani Organics,


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)