i want to calibrate a transmitter in the range of -100 to
300 mmwc. How it is possible to measure -100 to 300?.
Answers were Sorted based on User's Feedback
Answer / carl_ellis
Atmospherice pressure equals zero pressure (gauge).
Vacuum pressures are less than atmospheric so are negative with respect
to atmospheric pressure. -100mm would be a vacuum pressure.
Pressures above atmospheric are positive with respect to atmospheric
pressure.
Carl Ellis
Measure First
| Is This Answer Correct ? | 1 Yes | 0 No |
Answer / carl_ellis
-100mm wc is a vacuum with respect to atmospheric pressure
as a reference.
Given the range, the pressure transmitter is likely a
differential pressure transmitter.
-100mm wc pressure can be produced with a vacuum source,
like a hand pump connected to the high side, with the low
side open to atmosphere.
Or 100mm wc positive pressure can be applied to the low
side, with the high side open to atmosphere:
DP = high side minus low side
DP = 0 minus 100 = -100mm wc
The URV, 300mm wc, is a calibrated with a pressure source
connected to the high side, with the low side open to the
atmosphere.
Carl Ellis
Measure First
| Is This Answer Correct ? | 0 Yes | 0 No |
Describe a time when you anticipated potential problems and developed preventive measures.
what happen if i increase height of displacer type LT.accuracy or sensitivity are affected.why?
what is the working principle of vortex flowmeter
what is commissioning.
capillary type tranmitter using remote seals and capillary tubes filled with liquid of known of known density so if you wanna callibrate it just close block valve just before capilary @ high side & connect druck apply it's calibrated range pressure as normal calibration ensure low side block valve must be closed & if you asking commissioning so this is not a DP type pressure tx so simple if you wanna put in servie this tx just open high side block valve & than low side block same way i do it but still transmitter showing -13 ma.just tell me what i want to do ?
what are the possible reasons if a control valve fails to open/close?
Can any body help to get Rsemulator 5000 free from internet if possible then give me any link or any help .
WHAT IS A SYMPTOM OF FAULT WITH THE SPAN OF AN INSTRUMENT? 1)ZERO-POINT AND 100% ARE CORRECT BUT NOT A MID-RANGE. 2)ZERO-POINT IS CORRECT ,BUT 1005 INPUT IS NOT GIVING 100% OUTPUT. 3)LINEARITY PRBS. 4)1005 O/p IS CORRECT BUT ZEO POINT IS WRONG.
what is interlock ckt? why it is necessary in industrial machines ?
what is difference between OP AMP and differential amplifier?
full process of gas turbin?
what is the deferance between LCN and UCN.
Civil Engineering (5086)
Mechanical Engineering (4453)
Electrical Engineering (16638)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)