what is the diff betwen calibration and system suitability?
Answers were Sorted based on User's Feedback
Answer / srinivasu & phani
Calibration is only performance check for instrument, where
as System Suitability is the criteria in which we look for
various parameters like analyst, reagents, system for a
particular analysis.
| Is This Answer Correct ? | 27 Yes | 1 No |
Answer / rahul kamble m.pharm
CALIBRATION is just to check whether the instrument is work
according to SOP procedure or not? it is done for
minimising the errors in the reading.
whereas SYSTEM SUITABILITY is process in we check various
parameters like analyst, its qualification, reagent for
whether they are AR or RS grade. its standard, calibration
of glasswares, instruments. LAB. CONDITIONS ETC.
| Is This Answer Correct ? | 11 Yes | 0 No |
Answer / bhushan
Calibration is only performance check for instrument, where
as System Suitability is the criteria in which we look for
various parameters like analyst, reagents, system for a
particular analysis
| Is This Answer Correct ? | 6 Yes | 1 No |
Calibration is only performance check for instrument, where
as System Suitability is the criteria in which we look for
various parameters like analyst, reagents, system for a
particular analysis.
along with that calibration is done on perodicali butyou
must run sst in every analysis
| Is This Answer Correct ? | 3 Yes | 1 No |
Answer / manoj k. harne
In calibration we check the performance of instrument only by using primary or reference standards. But, in case of system suitability we checks over all performance of the analytical method against chemical reagents used, analyst,glassware,environment conditions. But, both activity closely related with each other. If instrument is not calibrated well then system suit can not meet criteria.in short calibration is precision of instrument and system suit is precision of proposed/intended method/procedure for analysis.
| Is This Answer Correct ? | 2 Yes | 0 No |
calibration means whether the instrument properly working
or not, when compared to reference standard based on some
parameters. if it fails it is a quality impacting insident.
system suitability means it is use full for analyst,
reagents, columns. these are intended use.
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / ch s raju
calibration is the performance check of the instrument
where as system suitability is your system is suitable for
that method we follow some ich guidelines for system
suitability parameters like resolution,RRT etc
| Is This Answer Correct ? | 2 Yes | 3 No |
Answer / narasaiah
both are the same if calibration is ok system suitability
parameters pass.when the entire system configeration is in
good condition then only system suitability will pass other
wise not pass
| Is This Answer Correct ? | 2 Yes | 10 No |
which one is better separation in high pressure liquide chromatogram orlowpressure chromato gram in hplc
why we are using benzene, anyline in acetic anhdride assay titration method?
What is a base line?
In the gas chromatography what is used carrier gas and make up gas?
Why a1% value is used for some product ? What is the criteria for selection a1% value ?
What is difference between" LOD" AND "VACCUME LOD"
what is the source of ir spectroscopy
2 Answers CTX Life Sciences, Mylan,
what is the use of tap density apparatus? why we calculate it? what is the difference between the bulk density and tap density?
1 Answers Alchemist, Alembic, Apex,
Why we use caffeine for HPLC calibration in Detector wave ength accuracy test
60 Answers Arvind, Aurobindo, Cipla, Reddy Labs, Shasun Chemicals,
what is a drug?
How many clauses in ISO!
1.What is the difference between method validation and method verification 2.Which guidelines proposed to method transfer