How can v calibrate ftir ? Plz explain me
Answers were Sorted based on User's Feedback
Answer / a.sudhakar
FT-IR IS CALIBRATED BY USING POLYSTYRENE FILM
PUT THE POLYSTYRENE FILM INTO THE SAMPLE HOLDER
AND SCAN IT FIXED WAVE NUMBERS ARE OBTAINED THEY ARE
3060
2849
1942
1601
1583
1154
1028
ACCEPTANCE CRITERIA IS +/- 1 CM
RESOLUTION: THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 2870 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 2849.5 CM SHALL BE MORE THAN 0.33
THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 1589 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 1583 CM SHALL BE MORE THAN 0.08.
| Is This Answer Correct ? | 29 Yes | 1 No |
Answer / t n rao
FTIR IS CALIBRATED BY USING POLYSTIRENE FILM THICKNESS OF
0.04mm(AS PER USP)0.05 mm(AS PER IP. THE RESOLUTION IN
ABSORBANCE B/T 2849.5 AND 2870 IS 0.33, IN B/T 1583 AND
1587 IS 0.08.
FREEQUENCY : EVERY DAY BEFORE USING AND MAJOR MAINTENANCE
JOB.
| Is This Answer Correct ? | 8 Yes | 8 No |
why the sucrose used in calibration of polarimeter
Why sodium lamp is used in polarimeter?
1 Answers Aurobindo, Micro Labs, Mylan,
why phenolpthalin gives pink colour when we titrate aganist sodium hydroxcide & phenolpthalin gives colourless titrate aganist hydrochlric acid which is un burate
how to calibrate hplc & gc
define polar and non polar solvent
Why we use Caffeine in hplc calibration
How do we get end points and how many end points are possible for citric acid and di-acid not theorotically answer should be given practically.
how we can identify the impurity is coming below loq at transfering site?
What is blank titration of potentiometrically.
if you have given one product then how you determine the impurity in that?
Why caffeine as a reference standard used for calibration of HPLC?
11 Answers Gangwal Chemicals, Orchid,
for which product require to do content uniformity? what is limit of cu?