How can v calibrate ftir ? Plz explain me
Answers were Sorted based on User's Feedback
Answer / a.sudhakar
FT-IR IS CALIBRATED BY USING POLYSTYRENE FILM
PUT THE POLYSTYRENE FILM INTO THE SAMPLE HOLDER
AND SCAN IT FIXED WAVE NUMBERS ARE OBTAINED THEY ARE
3060
2849
1942
1601
1583
1154
1028
ACCEPTANCE CRITERIA IS +/- 1 CM
RESOLUTION: THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 2870 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 2849.5 CM SHALL BE MORE THAN 0.33
THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 1589 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 1583 CM SHALL BE MORE THAN 0.08.
| Is This Answer Correct ? | 29 Yes | 1 No |
Answer / t n rao
FTIR IS CALIBRATED BY USING POLYSTIRENE FILM THICKNESS OF
0.04mm(AS PER USP)0.05 mm(AS PER IP. THE RESOLUTION IN
ABSORBANCE B/T 2849.5 AND 2870 IS 0.33, IN B/T 1583 AND
1587 IS 0.08.
FREEQUENCY : EVERY DAY BEFORE USING AND MAJOR MAINTENANCE
JOB.
| Is This Answer Correct ? | 8 Yes | 8 No |
If we have 5 strength which is not dose proportinate and excipients also diffrent in each strength then how we can proceed for Force degradation? and excipient are same but not dose propotinate the how FD?
what is the difference between polarimeter lamp and ir lamp
Why 10%,20%,30% Sucrose solution is used for the calibration of polarimeter?
5 Answers Cipla, LGC Promochem,
why salisylic acid not using now days for disdolution calibration?
Dear sir, This is R.Kadambavadani working in a pharma industry. while doing RS analysis of ibuprofen API, unknown peak eluting some times in the 0.6 Min RT. Kindly send me what is the exact reason behind that. also am using ACN the mobile phase.
why SGF media used for dissolution . What is the use of it .can we use other media of lower pH instead ot this?
In Number of Theoretical Plates (N) = 16 (tr/w)2. where tr: retention time, and W: peak width. My Question is here in formula what is 16.Can u explain briefly? ---Veerabhadrarao.M
3 Answers Micro Labs, Richter Themis, Sun Pharma,
how PH an important role in HPLC separation
why using the ph 4,7,9.2?
In HPLC Calibration, On which basis RSD Limit of noise test is fixed (NMT 33.0 % )
How can we calculate offset value pH measurment
How to calculate the signal to noise ratio.