How can v calibrate ftir ? Plz explain me
Answers were Sorted based on User's Feedback
Answer / a.sudhakar
FT-IR IS CALIBRATED BY USING POLYSTYRENE FILM
PUT THE POLYSTYRENE FILM INTO THE SAMPLE HOLDER
AND SCAN IT FIXED WAVE NUMBERS ARE OBTAINED THEY ARE
3060
2849
1942
1601
1583
1154
1028
ACCEPTANCE CRITERIA IS +/- 1 CM
RESOLUTION: THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 2870 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 2849.5 CM SHALL BE MORE THAN 0.33
THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 1589 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 1583 CM SHALL BE MORE THAN 0.08.
| Is This Answer Correct ? | 29 Yes | 1 No |
Answer / t n rao
FTIR IS CALIBRATED BY USING POLYSTIRENE FILM THICKNESS OF
0.04mm(AS PER USP)0.05 mm(AS PER IP. THE RESOLUTION IN
ABSORBANCE B/T 2849.5 AND 2870 IS 0.33, IN B/T 1583 AND
1587 IS 0.08.
FREEQUENCY : EVERY DAY BEFORE USING AND MAJOR MAINTENANCE
JOB.
| Is This Answer Correct ? | 8 Yes | 8 No |
about analytical chemistry
differance between L1 and L7 columns
How to define samples weight in moisture content
various methos to calculate precision in validation methods in HPLC?
What is the difference between the optics of UV and PDA Detectors ?
what is a drug?
what is end capping in hplc coulmn
2 Answers Lupin, Zydus Cadila,
What is difference between Chromatographic purity and Related substances analysis?
how you fix the limits of impurities?
What is impurities in pharma analysis?
Dear sir, This is R.Kadambavadani working in a pharma industry. while doing RS analysis of ibuprofen API, unknown peak eluting some times in the 0.6 Min RT. Kindly send me what is the exact reason behind that. also am using ACN the mobile phase.
what is the purge flow & how to calculate