How can v calibrate ftir ? Plz explain me
Answers were Sorted based on User's Feedback
Answer / a.sudhakar
FT-IR IS CALIBRATED BY USING POLYSTYRENE FILM
PUT THE POLYSTYRENE FILM INTO THE SAMPLE HOLDER
AND SCAN IT FIXED WAVE NUMBERS ARE OBTAINED THEY ARE
3060
2849
1942
1601
1583
1154
1028
ACCEPTANCE CRITERIA IS +/- 1 CM
RESOLUTION: THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 2870 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 2849.5 CM SHALL BE MORE THAN 0.33
THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 1589 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 1583 CM SHALL BE MORE THAN 0.08.
| Is This Answer Correct ? | 29 Yes | 1 No |
Answer / t n rao
FTIR IS CALIBRATED BY USING POLYSTIRENE FILM THICKNESS OF
0.04mm(AS PER USP)0.05 mm(AS PER IP. THE RESOLUTION IN
ABSORBANCE B/T 2849.5 AND 2870 IS 0.33, IN B/T 1583 AND
1587 IS 0.08.
FREEQUENCY : EVERY DAY BEFORE USING AND MAJOR MAINTENANCE
JOB.
| Is This Answer Correct ? | 8 Yes | 8 No |
please explian me about area normalisation method,diluted standard method,impurity vs impurity and which one has to be selected in method development
if the flow rate is increases then any changes in retention time in hplc
about method developement in hplc
Why only Copper standard is used to calibrate Atomic Absorption spectrophotometer?
Why do we use 500ppb sucrose and 500ppb 1,4 benzoquinone in TOC calibration?
What is difference between Potency & purity
How many clauses in ISO!
What are the differences between GC and HPLC?
why we use 0.1% acetone in gradient calibration of HPLC?
3 Answers Cipla, Dr Reddys, FDC,
why salicylic acid not use for dissolution.
2 Answers Cipla, Indo Baijin Chemicals,
what is pK & pKa value
How can we calculate offset value pH measurment