How can v calibrate ftir ? Plz explain me
Answers were Sorted based on User's Feedback
Answer / a.sudhakar
FT-IR IS CALIBRATED BY USING POLYSTYRENE FILM
PUT THE POLYSTYRENE FILM INTO THE SAMPLE HOLDER
AND SCAN IT FIXED WAVE NUMBERS ARE OBTAINED THEY ARE
3060
2849
1942
1601
1583
1154
1028
ACCEPTANCE CRITERIA IS +/- 1 CM
RESOLUTION: THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 2870 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 2849.5 CM SHALL BE MORE THAN 0.33
THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 1589 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 1583 CM SHALL BE MORE THAN 0.08.
| Is This Answer Correct ? | 29 Yes | 1 No |
Answer / t n rao
FTIR IS CALIBRATED BY USING POLYSTIRENE FILM THICKNESS OF
0.04mm(AS PER USP)0.05 mm(AS PER IP. THE RESOLUTION IN
ABSORBANCE B/T 2849.5 AND 2870 IS 0.33, IN B/T 1583 AND
1587 IS 0.08.
FREEQUENCY : EVERY DAY BEFORE USING AND MAJOR MAINTENANCE
JOB.
| Is This Answer Correct ? | 8 Yes | 8 No |
what is a differance between assay by HPLC and assay by titration?
before starting analytical method valodation what you checking? and how giving preference to start validation?
If i performing one analysis on C8column then the same analysis done on C18 column then what will happen about Rt;Resulution;Peak shape Explain ?
what is difference between GC & GC-HS
why potassium dicromate use in calibration of UV VIIBLE SPECTROCOPY ?
What is calibration procedure of gredient HPLC pump
What is AV value in test uniformity content ,describe calculation
why we get negative values in uv when i placed sample which has some concetration but not in standard
what is the difference between mcg and ppm?
what is impurity profile. how to interpret this impurity profile to a drug product or drug substance.
which light is used in FT-IR spectroscopy
Why we more noise in PDA detector compared to UV detector?