Answer / a.sudhakar
FT-IR IS CALIBRATED BY USING POLYSTYRENE FILM
PUT THE POLYSTYRENE FILM INTO THE SAMPLE HOLDER
AND SCAN IT FIXED WAVE NUMBERS ARE OBTAINED THEY ARE
3060
2849
1942
1601
1583
1154
1028
ACCEPTANCE CRITERIA IS +/- 1 CM
RESOLUTION: THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 2870 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 2849.5 CM SHALL BE MORE THAN 0.33
THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 1589 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 1583 CM SHALL BE MORE THAN 0.08.
| Is This Answer Correct ? | 29 Yes | 1 No |
Answer / t n rao
FTIR IS CALIBRATED BY USING POLYSTIRENE FILM THICKNESS OF
0.04mm(AS PER USP)0.05 mm(AS PER IP. THE RESOLUTION IN
ABSORBANCE B/T 2849.5 AND 2870 IS 0.33, IN B/T 1583 AND
1587 IS 0.08.
FREEQUENCY : EVERY DAY BEFORE USING AND MAJOR MAINTENANCE
JOB.
| Is This Answer Correct ? | 8 Yes | 8 No |
Hello Dear all; I have a technical question related to analytical Chemistry, I have a sample of (40% w/w) Ferric Chloride solution (for water treatment purposes which consist a little portions of Ferrous Chloride and Hydrochloric acid. I need to know the methods to determine the percentages of each Ferrous Chloride and Chloride acid. Also I want to make sure that this solution is truly (40% w/w). So ANY CLUE !!! Thanks in advance.
what is mean by 21 CFR PART? In that what is 21 ?
How to determine water content of thioanisole? which interfere by KF titration?
what is specificity?
4 Answers Amneal Pharmaceuticals, Aurobindo,
HPLC Calibration was done in Reverse Phase, but not in normal phase why explain?
If we change lenght of ph eletrode wire then it will affect on ph ?
why we are use metthanol variously in kfr?why don't use other solutions?
How to check hplc detecter sensitivity?
How to decided test concentration in GC analysis.
how can we decide the limit of cleaning validation? What is the criteria for it?
Why are using in Hi,zero air, Ni gass are used GC
5 Answers Aurobindo, ELBS, Vimta Labs,
How to calculate the signal to noise ratio.