How can v calibrate ftir ? Plz explain me
Answers were Sorted based on User's Feedback
Answer / a.sudhakar
FT-IR IS CALIBRATED BY USING POLYSTYRENE FILM
PUT THE POLYSTYRENE FILM INTO THE SAMPLE HOLDER
AND SCAN IT FIXED WAVE NUMBERS ARE OBTAINED THEY ARE
3060
2849
1942
1601
1583
1154
1028
ACCEPTANCE CRITERIA IS +/- 1 CM
RESOLUTION: THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 2870 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 2849.5 CM SHALL BE MORE THAN 0.33
THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 1589 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 1583 CM SHALL BE MORE THAN 0.08.
Is This Answer Correct ? | 29 Yes | 1 No |
Answer / t n rao
FTIR IS CALIBRATED BY USING POLYSTIRENE FILM THICKNESS OF
0.04mm(AS PER USP)0.05 mm(AS PER IP. THE RESOLUTION IN
ABSORBANCE B/T 2849.5 AND 2870 IS 0.33, IN B/T 1583 AND
1587 IS 0.08.
FREEQUENCY : EVERY DAY BEFORE USING AND MAJOR MAINTENANCE
JOB.
Is This Answer Correct ? | 8 Yes | 8 No |
about method developement in hplc
We use hicrome 10 cxs column in hplc analysis....pls tell me.. what is the meaning of 10 cxs?
What is AV value in test uniformity content ,describe calculation
Why irmoisture balance is used for corrugated box moisture testing
If change in specification which parameter require to do for validation? if change in chromtographic condition then which parameter? if api change then which parameter? if change composition then which parameter? if old method not work out then whicj parameter? if additional one impurity added then which parameter of validation require to do on above each conditions? elaborate separately
What is the formula for relative diffrence for standard solution in solution stability in validation?
how do start hplc method development? how much sample required for developing the method development?
3 Answers BMS, Cipla, Indoco Remedies, Sun Pharma,
Why Holium cholorate used in UV Spectro photometer?
why hexadecane is used in calibration of FID IN GC THERE IS NO REMAINING SOLVENTS
How to decided test concentration in GC analysis.
What r the limitatoins of uv visible spectroscopy ? Why Water use in blank standard for uv visible spectroscopy ?
How much is the standard area for glc analysis