In my analysis by HPLC,In sample main peak and placebo
peak are eluting with shorter resolution ( almost look like
doublet). how can i get max resolution ? Is changing
gradient method is usefull?
Answers were Sorted based on User's Feedback
Answer / wasim shaikh
You can try one or more of the following:
1) Changing the column type (c8 to c18, etc.)
2) Changing the Mobile phase composition (increase buffer
etc.)
3) Decrease the particle size on column (from 5 to 4, 3
etc.)
4) Increase the column length
5) Decrease the flow rate
6) If all of the above fails, then as you suggested,
gradient method can be used to separate the two peaks.
P.S. : Gradient method can also be used as an alternative
to the above methods and not necessarily the ultimate
solution.
| Is This Answer Correct ? | 32 Yes | 1 No |
Answer / naveen
U CAN DECREASE THE FLOW RATE
AND INJECTION VOLUME
| Is This Answer Correct ? | 1 Yes | 1 No |
Answer / gaurav
You can increase the injection volume.Changing gradient
method is not required.
| Is This Answer Correct ? | 2 Yes | 14 No |
why lactose has less charges when packed in steel containers when compared with packing of lactose in polyamide material?
what is the general mechanism of high performance liqiud chromatography?
7 Answers Ajanta Pharma, Amneal Pharmaceuticals,
if tech transfer fails whay require to do?
why sucrose not used in polarimeter calibration?
What is delay volume?
How to fix the concentration for RS by HPLC
control of resolution in calibration of IR
in HPLC injection volume is increase ,R.T will increase ?
5 Answers Claris Lifesciences,
if you have given a blank water how you analyse by general analytical techniques?
Why dissolution test apparatus calibration with salycilic acid tablets was stopped by FDA&USP from the year of 2009?
Which std. used for standrdisation of ZnSO4?
what is rs test why we are perfoming rs test