In my analysis by HPLC,In sample main peak and placebo
peak are eluting with shorter resolution ( almost look like
doublet). how can i get max resolution ? Is changing
gradient method is usefull?
Answers were Sorted based on User's Feedback
Answer / wasim shaikh
You can try one or more of the following:
1) Changing the column type (c8 to c18, etc.)
2) Changing the Mobile phase composition (increase buffer
etc.)
3) Decrease the particle size on column (from 5 to 4, 3
etc.)
4) Increase the column length
5) Decrease the flow rate
6) If all of the above fails, then as you suggested,
gradient method can be used to separate the two peaks.
P.S. : Gradient method can also be used as an alternative
to the above methods and not necessarily the ultimate
solution.
| Is This Answer Correct ? | 32 Yes | 1 No |
Answer / naveen
U CAN DECREASE THE FLOW RATE
AND INJECTION VOLUME
| Is This Answer Correct ? | 1 Yes | 1 No |
Answer / gaurav
You can increase the injection volume.Changing gradient
method is not required.
| Is This Answer Correct ? | 2 Yes | 14 No |
how to calibrate AAS ? give me the procedure?
verification of wave number why you are use in IR calibration
WILL YOU DRY DST BEFORE FACTOR DETERMINATION BY USING KARL FISHER TITRATOR
4 Answers Aurobindo, Meghmani Organics,
Why use in sodium salicylate in uv lamps calibration
1 Answers Auctus Pharma, Reddy Labs,
from where i get mortar pastle for glass bottle crush? it's required for testing of it?
for heavy metal test lead used in which form pure form or any other form for preparation of lead standard sol
1 Answers Watson Pharmaceuticals,
what is the relation of normality & molarity ? give me formula?
In Number of Theoretical Plates (N) = 16 (tr/w)2. where tr: retention time, and W: peak width. My Question is here in formula what is 16.Can u explain briefly? ---Veerabhadrarao.M
3 Answers Micro Labs, Richter Themis, Sun Pharma,
in hplc chromatogram started from left to right and in uv spectrum started from left to right
How to compare XRD graphs against standard and carry polymorpism study of API's by powder XRD method?
What is Signal to noise ratio?
What the difference in loss on drying and residue on ignition