In my analysis by HPLC,In sample main peak and placebo
peak are eluting with shorter resolution ( almost look like
doublet). how can i get max resolution ? Is changing
gradient method is usefull?
Answers were Sorted based on User's Feedback
Answer / wasim shaikh
You can try one or more of the following:
1) Changing the column type (c8 to c18, etc.)
2) Changing the Mobile phase composition (increase buffer
etc.)
3) Decrease the particle size on column (from 5 to 4, 3
etc.)
4) Increase the column length
5) Decrease the flow rate
6) If all of the above fails, then as you suggested,
gradient method can be used to separate the two peaks.
P.S. : Gradient method can also be used as an alternative
to the above methods and not necessarily the ultimate
solution.
| Is This Answer Correct ? | 32 Yes | 1 No |
Answer / naveen
U CAN DECREASE THE FLOW RATE
AND INJECTION VOLUME
| Is This Answer Correct ? | 1 Yes | 1 No |
Answer / gaurav
You can increase the injection volume.Changing gradient
method is not required.
| Is This Answer Correct ? | 2 Yes | 14 No |
WHY KBr is used in IR Spectroscopy
which solvent is highest poler acetonitrile dimethyl formamide methanole tetrahydrofurane buffer solution
How you collect dissolution media while in stirring condition of a basket or a paddle?
Why dissolution apparatus having 8 jars ?
In standard weight box used for calibration of analytical balance there are two numbers (Qty) of 20mg, 200mg,2.oGM weights. Why?
What is the formula for Tapdensity /
6 Answers Dr Reddys, GR Intrachem,
What is the difference between Paddle and Basket Why we are using Basket or paddle in dissolution.
7 Answers Apotex, Cipla, GSK, Intas, Zydus Cadila,
How to determine retention time in Chromatography
how the compound separate in coulunm,explain
please explain about aluminium hydroxide assay
why we are using 6 dosage units in dissolution, and why we are using 6 tablets in disintegation test.is there any scientific reason beyond 6?
could negative ions be produced by bombardment process in mass spectrometry?