Why do we use 500ppb sucrose and 500ppb 1,4 benzoquinone in
TOC calibration?
Answers were Sorted based on User's Feedback
Answer / vishal shah
In TOC calibration 1,4 Benzoquinone used because 1,4
Benzoquinone is tufest to oxidized and sucrose is easly
oxidized so cover upper and lower range for calibration.
| Is This Answer Correct ? | 53 Yes | 0 No |
Answer / dev rawat
In Toc Analyzer Calibration (System suitability) we are preparing Standard sucrose of 500PPb and Benzoquinone Preparing of 500ppb . Sucrose is easy to oxidized and p- Benzoquinone is hard to decompose(oxidized).Its molecule bound not easy to oxidized so the response efficiency defined in USP (united state pharmacopoeia )its response efficiency not less then 85% and not more then 115% . BQ run as control sample in the References of Sucrose .And Its Calculation methods [(rss-rw ) / (rs-rw)]100.
| Is This Answer Correct ? | 15 Yes | 0 No |
Answer / amarjeet kumar
Any sample in TOC analyzer is being purged with inert gas aftrr mixed with an acid and 1,4 benzoquinone is the toughest as said above to get oxidised others may get easily oxidised.
| Is This Answer Correct ? | 9 Yes | 0 No |
Answer / dignesh udwadia
sucrose is high oxidation and 1,4-benzoquinone is vary low oxidation to cover higher and lower oxidation range we are used above mention material for calibration.
| Is This Answer Correct ? | 7 Yes | 1 No |
Answer / umamahesh
sucrose is high oxidation and 1.4- benzoquene is very low oxidation to cover higher and lower range
| Is This Answer Correct ? | 3 Yes | 0 No |
Answer / venkatesh palla
1.In the TOC Calibration Sucrose working as a test solution
and 1,4 benzoquinone working as a standard or Control
solution.
| Is This Answer Correct ? | 13 Yes | 21 No |
What is the difference between Spectroscopy and spectrophotometry,What are the applications?
4 Answers Caplin Point, Cipla,
why IR graph or spectra comes down ? Any reason
signifcance of detector lenearity in calibration f hplc
In C8,C18....etc have 4&5 etc... microns Columns is there like that how to separate Chiral columns as per Micron size?
what is the use of tap density apparatus? why we calculate it? what is the difference between the bulk density and tap density?
1 Answers Alchemist, Alembic, Apex,
why use k2cr2o7 , kcl h2so4 in uv calibration
Which is the best mobile phase to separate organic acids using C8 column. And how should the samples be prepared for hplc?
Related substance method equivalency on control sample or spiked sample?
Wt s the use of linearity in HPLC method validation
What is FDA form 482c?
If i performing one analysis on C8column then the same analysis done on C18 column then what will happen about Rt;Resulution;Peak shape Explain ?
Why dissolution test apparatus calibration with salycilic acid tablets was stopped by FDA&USP from the year of 2009?