How to calibrate vertex flow meter?
How to calibrate mass flow meter?
Vortex:
Practically it is a volumeteric flow measuring device and it is not possible to calibrate online in field in most of the cases (Provide you have a proper bypass arrangement and proovers in place) , hence it is to be sent to calibration LAB where it will be compared with MASTER meter.
This MASTER meter must have superior accuracy than UUC (Unit Under Calibration).
Usually the large volume flow is tough to establish at calibration LAB hence Gravymeteric method is used in which water is passed through UUC and compared with high accuracy Load-cell. Since water density is known and constant, volume flow is calculated and new K-factor of meter is calculated.
Mass-flowmeter / Coriolis meter : Same Above mentioned Gravymeteric method is used.
Hope this will help ..!!!
Jigar Anadkat
Is This Answer Correct ? | 0 Yes | 0 No |
When can a bypass not used on a positioner?
What are the Components of Siemens S7 300 Series PLC system ?
Can please tell mi about SIOS, PRM, SERVER, and what is the diference between PIN LAN, PCS LAN,APC Server, and please send the Doc if possible for referance,mail id patil_amit17@yahoo.com
what is zenier barier and what means zone 1,zone 2,and zone 3
2 Answers Bapco Bahrain Petroleum Company,
Flow tax range 0 to 100 mmh2o ,flow rate 40 tons /hr, I want change flow rate example 42 tons/hr so tx range ?
What is the meaning of the term "g force"?
3 Answers Reliance, Saint Gobain, Thermax,
What is use of condensate pot
5 Answers Alstom, Petrocil, Samsung,
Can all intrinsic safe transmitters are provided with inbuilt zener barriers? why
How to Calibrate Transducers? Can you give me a step by ste procedure?
what is the difference between venturi and orifice meters?
what do you mean by cathodic protection?
Tell me the field calibration procedure step by step with diagram of temperature transmitter, temperature indicator, flow indicator, level indicator, control valve, thermocouple, pressure indicator (digital & analogue).