iam using ph instrument lab india make.3ponit and 5point.in
calibration time 3point slope will come 98.89and 96.87.that
limit is 85 to 105.what basis slope will come.any
calucalation between them.5point wiicome slope 99.89?why
3p0int came 2 values
Answer / firoj m. mulani
It is simple,we know pH is inversally praportional to the Mv
ie If pH is Increases mv is decreases.
then,for calculating the slope at 5 point a simple formula
as below,
Mv1.68 - Mv4.01*100
---------------------
pH4.01 - pH1.68*59.16
Mv4.01 - Mv6.87*100
---------------------
pH6.87 - pH4.01*59.16
for next point's,formula as above
if 5 point pH it gives 4 values came of slope.
| Is This Answer Correct ? | 1 Yes | 0 No |
0.1 atom gram of the unknown element( X)reacts with 2.4 gram of O2 to produce the oxid product . can you tell us, what is the oxid formula ? a) XO b)X2O3 c)XO2 D) X2O4
what is dissolution
If inhouse hplc related substance method is completly diffrent from Usp for finished proďuct with diffrent impurities then how require to prove method equivalecy?
Why water not directed by FID detector
2 Answers Amoli Organics, Lupin,
Which type of material analysis in GC using Theramal conditivity detector ?
in hplc calibration we are using 0.05% acetone why we are using that one and what happen when its conc increases or decreases
4 Answers Aurobindo, Cadila, Martin,
What is the accrptance criteria in RSD for RS method precision on basis of impurity percentages?
what is lod and loq ?,why use k2cr2o7 , kcl h2so4 in uv calibration ?,why use benzophenone & caffene acetone in hplc calibration ?,what is leading peak in hplc ?why we do the calibration of limit of stry light in hplc & uv ?
why required specificity in valdation
why using factor 1.00639 in burette calibration of kf
what is specificity?
4 Answers Amneal Pharmaceuticals, Aurobindo,
Which parameter require to do for analytical method equivalency?