why we are used 4,7,9,2 buffers for the ph calibration
Answers were Sorted based on User's Feedback
Answer / pharmainterview
Basic concept is you need to check in acidic,neutral and
basic range. By covering all these regions you can check
anyware.
As per usp the difference between two adjacent buffers
should not more than 4.
Regards
pharmainterviewquestions.blogspot.in
| Is This Answer Correct ? | 41 Yes | 0 No |
Answer / amarsingh
Because by using this all pH buffer we can cover all range
of pH (Acidic , Neutral & Basic ) So we use this range
| Is This Answer Correct ? | 16 Yes | 4 No |
Answer / pankaj shrivastava
Calibration points should cover required ph adjustment of
solution. if you use only 2 to 7 ph range for ph adjustment
then you can calibrate the ph with 2.4.7 ph.
| Is This Answer Correct ? | 2 Yes | 1 No |
Answer / arash babakhanian
pH is determined from this equation in the pH meter
pH = [(Eobs-K)/0.059]n
Eobs= observed potential
K= constant paramere
0.059=slope
constant parameters k and 0.059 should be calibrated. In
fact two acidic buffers and two alkalic buffers are used to
fix slope and K constant . In this case, the relation
between Eobs and pH is analyticaly change
| Is This Answer Correct ? | 1 Yes | 2 No |
What is the difference between purge septum flow and column flow in gas chromatography?
why did uyou choose analytical chemistry?
In which situation we require to prepare the standard solution from sample in Related substance method?
What are the concentration limit for class-1 residual solvents in ICH ?
can some other acids like nitric acid or HCl can be used in the calibration of UV.Vis. spectrophotometer????
In lcms analysis some times we get sodium and potassium adducts. if the compound is havine sodium ions how can we confirm wether sodium is from sample or it is a adduct?
2 Answers Teva Pharmaceuticals,
formula for %RSD.
What is Delay & Dead Volume
What is the diference between residual solvents and organic volatile matter
Difference Between End Cap HPLC Column and Non End Cap Column
How to check hplc detecter sensitivity?
on saturation solubility study data how we can find out the bcs class of drug?