what is the criteria for choosing capillary column in GC??is it dependent on sample load?
Answers were Sorted based on User's Feedback
capillary column range of the 400 temp but it length of 30
meter and packed column length is 3 meter otherwise choose
the acuratly & high sensitivitis peak the sample of minor
impurities collect the column then choose the capillary column.
not a dependent on the sample load.
Is This Answer Correct ? | 3 Yes | 1 No |
Answer / kanakadurga
Packed columns have higher sample capacity than capillary
column.
Improvements in detector sensitivity have also reduced the
need for large samples.For almost all other samples,
capillaries provide much better efficiency (narrow peaks)
which leads to greatly improved peak separation. Capillary
columns are highly sensitive and efficient. these are two
types WCOT and PLOT.
Yes these are depends on Sample Load. Very Low quanity of
samples are enough for the analysis.
Is This Answer Correct ? | 2 Yes | 2 No |
Answer / sachin patil
Because of the peak shape of solvent observe broad and high consumption of gas flow require in analysis through pack column. Its also need more concentration of sample require than capillary column analysis.
Is This Answer Correct ? | 0 Yes | 0 No |
in api coa contains only process impurities and in product coa contains degradation impurities?
How to calculate coreletion coefficient
How to do regeneration of Metacarb Pb plus column?
wen v have polr solvent then which type of column use in hplc
What is the procedure for UV-spectrophotometer calibration?
1)why use benzophenon,uracil in hplc calibration. 2)why use sucrose in polarimeter calibration. 3)what is S/N ratio, noise and drift in hplc 4)how we can find lod and loq in method validation
Please give idea about method development for cleaning method and how maco level establish
How to interpret XRD graphs against standard and limit for same?
what is the difference betwen the analytical column & peparative column?
1.What is the difference between method validation and method verification 2.Which guidelines proposed to method transfer
What is rs test why we are performing rs test?