What are the problems faced while testing HART instruments?
Answer / carl_ellis
Failure to connect with HART can be a laundry list of
issues, the most common of which is the
lack of minimum 230 ohm resistance in the loop.
Other reasons can be
- HART probes connected across the DC power supply, not in
the loop
- failed DC power supply
- electrically noisy DC power supply
- HART modem lacks installed PC driver or internal setup
(serial modems were touchy)
- HART device not at expected HART address (address zero for
point-to-point comm)
- device does not have HART option or HART option has been
disabled.
- HART option replaced by different fieldbus comm card (it's
happened! !)
- Connecting across the test terminals on a pressure
transmitter where there's no HART signal
- HART communicator is dysfunctional or its batteries are
discharged
- lack of DD in communicator/program limits access to the
common commands
Carl Ellis
Measure First
Is This Answer Correct ? | 1 Yes | 0 No |
Where Can i take a SPI INtools training in Mumbai? In Rolta India Ltd to whom I should approach regarding of SPI INtools Training?
What is difference between reset and initialization in industrial automation?
Why do you want to work here
What is the minmum distance between the electrical panel and Instrument panel
what is compensation? temprature & pressure
what is mean by loop checking?
What are the interlocks we have to check during the start up of LP/HP compressor?
What is the formula of vabration sesore. i mean how to covert mils to volt plz share if any body knows very argent : my mai is merajalam143545@yahoo.com
Is wet leg type system nessesary to measure water level of steam DRUM? How the transmitter is calibrated to measure this level?
how we can check level probe and controller
What is difference between profibus and modbus communication
what is diff between the motor and pump