What are the problems faced while testing HART instruments?
Answer / carl_ellis
Failure to connect with HART can be a laundry list of
issues, the most common of which is the
lack of minimum 230 ohm resistance in the loop.
Other reasons can be
- HART probes connected across the DC power supply, not in
the loop
- failed DC power supply
- electrically noisy DC power supply
- HART modem lacks installed PC driver or internal setup
(serial modems were touchy)
- HART device not at expected HART address (address zero for
point-to-point comm)
- device does not have HART option or HART option has been
disabled.
- HART option replaced by different fieldbus comm card (it's
happened! !)
- Connecting across the test terminals on a pressure
transmitter where there's no HART signal
- HART communicator is dysfunctional or its batteries are
discharged
- lack of DD in communicator/program limits access to the
common commands
Carl Ellis
Measure First
Is This Answer Correct ? | 1 Yes | 0 No |
a pressure transmitter with a range of 0 to 2500kpag has a specified accuracy of 0.5 % of span.It is to be calibrated using a comparison gauge to have an accuracy of 0.1% of span.The comparison gauge has a range of 0 to 3000kpag What is the accuracy ratio? is this ratio acceptable?
1)why we are using mostly Ultrasonic Flow meters to measure flow, rather than Radar waves? 2)why we are using Ultrasonic waves and Radar waves to measure the Level?
give me the formula for analog to digital converter and also d to a converter
What is the exact output of eddy current proximity probe? For vibration mesurment?
How do you troubleshoot accelerometers used in Vibration monitoring/tripping
what is difference btween rotameter and orifice plate?
If a control valve is hunting means means if we are giving 50% to the cv, the cv will open at 51% then slowly slwoly it will reaches 49% , so what to do for solve this problem. The cv is double acting and mounted for anti surge operation, it having 4 nos. volume booster, one sov, one positioner, etc..?
hay friends, i am in trouble to understand PID controller. several times i have study but it is not clear to me have you a simple way to understand.
what is the diffrence between LCN and UCN
Explain signal isolator
How will u convert the mmh2o in mass or Volumetic flow rate in DP Flow Transmitter. Give me the Brief explain using the formula
1 Answers Walchandnagar Industries,
How to calibrate Field bus transmitter? procedure ?