Answer / karthigai selvan
in the field have to isolate first hp and lp side isolate
valves
fill the water on both side of both impulse tubes same time
both tubes should be in ambient(open to atmosphere) check
zero if not do zero trim then line up
thanks
| Is This Answer Correct ? | 9 Yes | 4 No |
Answer / sahnis
First step is to isolate the trasmitter chamber from tank.
Close HP first and then LP. After that drain the Chamber by
opening drain valve on the bottom of the chamber. Then
connect a trasparent hose and keep it parallel to the
chamber to know the level inside the Transmitter chamber.
Then start filling the Fluid untill the low level. When the
low level reached, the zero adjustment( 4mA output from tx)
can be done by turning the zero adjustment screw. Then fill
the fluid upto highest level and turn the span to get 20mA.
Calibration is done and the instrument is now ready to be on
loop.
| Is This Answer Correct ? | 2 Yes | 3 No |
what are the steps to follow force a particular input or out put from ics triplex esd system?
how to set calibration range in a DP transmitter to measure level for a closed tank and for an open tank,and also please explain me how to set calibration range for a DP flow transmitter in a line ?
How to rerange a dp type level transmitter after installation?
What are primary elements of measuring pressure? Which type of pressure can be measured by these elements?
on interface level transmitter model(RTG3302-EMERSON) What is threshold valve how can we select correct threshold value T1,T2,T3,T4 When interface water & hydro carbon
WHAT IS SELF CLEAN ELCTRODE IN MAGNETIC FLOWMETER?? WHAT DO YOU MEAN BY THIS TERM??
With N nodes and B branches what will have a network ???
why use 50% lel methane gas concentration for gas detector calibration instead of 60% lel gas concentration whereas the hi alarm set at 60% lel.
I want to use 1/2 inch SS316 tube in my high pressure & temperature (64 kg/cm2 & 500 deg C)after condensate pot in pressure tranmistter. 1) Is it possible to use 1/2" ss316 tube in high pressure and temperature line. Tube manufacturer is local vendor. 2)On sarting the plant, We need to do system fluse that time condesate is not in the condesate pot so steam is directly contact with the tube. We have no exact calculation to know our tube is ok or nor at this time. pleae,reply.
Flow tax range 0 to 100 mmh2o ,flow rate 40 tons /hr, I want change flow rate example 42 tons/hr so tx range ?
How to measure density of oil in your kitchen. i.e indirectly to verify the quality of oil using min. gadgets available in the kitchen/house.
Why mixer nd grinder Is example of constant torque load not variable torque load ???
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)