why a ph meter require KCL solution for their storing
condition?
Answer / dr. a. bera, cdl
When not in use, the glass probe tip must be kept wet at all times to avoid the pH sensing membrane dehydration and the subsequent dysfunction of the electrode.
A glass electrode alone (i.e., without combined reference electrode) is typically stored immersed in an acidic solution of around pH 3.0. In an emergency, acidified tap water can be used, but distilled or deionised water must never be used for longer-term probe storage as the relatively ionless water "sucks" ions out of the probe membrane through diffusion, which degrades it.
Combined electrodes (glass membrane + reference electrode) are better stored immersed in the bridge electrolyte (often KCl 3 M) to avoid the diffusion of the electrolyte (KCl) out of the liquid junction.
| Is This Answer Correct ? | 72 Yes | 5 No |
what is the difference between BDS and ODS Columens and both are same or not
2 Answers Intas, IPQA, Micro Labs, Vasava Engineering Private Limited,
In gas chromatographic analysis sometimes we use headspace, why it is required instead of manually injection?
what is the general mechanism of high performance liqiud chromatography?
7 Answers Ajanta Pharma, Amneal Pharmaceuticals,
(0.41)d what is d?
How do you give training to a new chemist
why we use TLC?
why we are used 4,7,9,2 buffers for the ph calibration
4 Answers Aurobindo, Enaltec Labs,
what is the diffrence between absorbance and absorption in the beer,s lamber formula ?
WHAT IS THE PARAMETER FOR THE SELECTION OF MOBILE PHASE IN HPLC?
What is use of LOD and LOQ in validtion and give range and in calculation why we use 3.3* LOD here why we use 3.3
2 Answers Dr Reddys, IPCA, Reddy Labs,
what is DIPE limit in residual solvent?
Why karl fisher factor is 5 ?