Answer / dr. a. bera, cdl
When not in use, the glass probe tip must be kept wet at all times to avoid the pH sensing membrane dehydration and the subsequent dysfunction of the electrode.
A glass electrode alone (i.e., without combined reference electrode) is typically stored immersed in an acidic solution of around pH 3.0. In an emergency, acidified tap water can be used, but distilled or deionised water must never be used for longer-term probe storage as the relatively ionless water "sucks" ions out of the probe membrane through diffusion, which degrades it.
Combined electrodes (glass membrane + reference electrode) are better stored immersed in the bridge electrolyte (often KCl 3 M) to avoid the diffusion of the electrolyte (KCl) out of the liquid junction.
| Is This Answer Correct ? | 72 Yes | 5 No |
What is use of LOD and LOQ in validtion and give range and in calculation why we use 3.3* LOD here why we use 3.3
2 Answers Dr Reddys, IPCA, Reddy Labs,
why should we perform dissolution PVT calibration only by UV spectrometer, not by HPLC ?
why we get only one peak in assay by hplc but in rs by hplc get more peaks
what is the identification test of volatile substance? what is volatile compound?
what parameters we will consider while developing a HPLC method and how we confirm our HPLC method is valid?
What is FDA form 482c?
why multimedia dissolution require to do?
what is the meaning of linearity,principle of co-relation coefficient?
why peak purity not considered in GC?
4 Answers DRL, Reddy Labs, Vimta Labs,
IN DISSOLUTION TEST S1AND S2 AND S3 CONDITIONS ALL ARE FAILED THEN WHAT IS THE DESSICION U CAN TAKE
What is the difference between Spectroscopy and spectrophotometry,What are the applications?
4 Answers Caplin Point, Cipla,
How to fix the sample concentration in related substances development.
3 Answers Novartis, Watson Pharmaceuticals,