why a ph meter require KCL solution for their storing
condition?
Answer / dr. a. bera, cdl
When not in use, the glass probe tip must be kept wet at all times to avoid the pH sensing membrane dehydration and the subsequent dysfunction of the electrode.
A glass electrode alone (i.e., without combined reference electrode) is typically stored immersed in an acidic solution of around pH 3.0. In an emergency, acidified tap water can be used, but distilled or deionised water must never be used for longer-term probe storage as the relatively ionless water "sucks" ions out of the probe membrane through diffusion, which degrades it.
Combined electrodes (glass membrane + reference electrode) are better stored immersed in the bridge electrolyte (often KCl 3 M) to avoid the diffusion of the electrolyte (KCl) out of the liquid junction.
| Is This Answer Correct ? | 72 Yes | 5 No |
What is the lab normality??
In standard weight box used for calibration of analytical balance there are two numbers (Qty) of 20mg, 200mg,2.oGM weights. Why?
From where 16 or 5.545 is derived in theoretical plates calculation formula by USP or JP/BP/DAB and what is meaning of this coefficient and impact in calculation ?
2 Answers Nectar Lifesciences,
what is the source of ir spectroscopy
2 Answers CTX Life Sciences, Mylan,
why we are used 4,7,9,2 buffers for the ph calibration
4 Answers Aurobindo, Enaltec Labs,
What is the deference between Residual solvents and OVI
What is the difference between BD
How to choose Normal phase or Reverse Phase HPLC method Conditions for given sample?
How you will select the runtime in HPLC for a drug in method development.
6 Answers Aurobindo, GVK, Lesanto Laboratories,
How can we calculates the solubility of the drug in different medias by injecting in to hplc????(solubility studies)
2 Answers Super Pharmaceuticals,
how can give the expiry period and restadardisation of volumetric solution
0 Answers Meenaxy Pharma, SRF,
Why base line not start from zero in GC Chromatography analysis.