why a ph meter require KCL solution for their storing
condition?
Answer / dr. a. bera, cdl
When not in use, the glass probe tip must be kept wet at all times to avoid the pH sensing membrane dehydration and the subsequent dysfunction of the electrode.
A glass electrode alone (i.e., without combined reference electrode) is typically stored immersed in an acidic solution of around pH 3.0. In an emergency, acidified tap water can be used, but distilled or deionised water must never be used for longer-term probe storage as the relatively ionless water "sucks" ions out of the probe membrane through diffusion, which degrades it.
Combined electrodes (glass membrane + reference electrode) are better stored immersed in the bridge electrolyte (often KCl 3 M) to avoid the diffusion of the electrolyte (KCl) out of the liquid junction.
Is This Answer Correct ? | 72 Yes | 5 No |
why hydrogen and zero air used for flame ignition in gc?
What are the limits for uniformity dosage content for tablets and capsules
why require the ph, buffer during hplc mobile phase?
only one end point is possible for citric acid and we get 2 end points for di acid why?
Ratio of Water and Acetonitrile(diluent) is not recommended to use in cleaning method validation,why?
0 Answers Aurobindo, Reddy Labs,
In HPLC calibration, caffeine is used as primary standard for wave length calibration due to caffeine is having dual maxima at 273 & 205 nm and one minima at 245 nm. Any body can give reference of these details from any pharmacopeia (with chapter no.) or any other authentic guideline?
why we need to do karl ficher fector ?
how do you established the RRF for unknown impurity
Can any one tell what is RECALL OF PROCEDURES is
what is the principle of UV Vis spectroscopy, AAS, ICP OES,ICPAES, ICP-MS and FTIR
What is the difference between known and unknown impurities?
Tell me about analytical method validation in QC