What is output of Thermocouple at 25 deg c
Answers were Sorted based on User's Feedback
Answer / b0nz0
It will depend on the type of thermocouple and the
temperature of the reference (cold) junction.
If you measure 25degC in an ambiemt temp of 25degC you
should read 0mV.
| Is This Answer Correct ? | 21 Yes | 1 No |
Answer / carl_ellis
Question: What is output of Thermocouple at 25 deg C?
I'd flunk all of you except B0nz0 because only B0nz0
understands thermocouple thermometry.
The answer is, you don't know because the cold junction
temperature is not specified. The question attempts to
find out if the concepts of a thermal gradient and cold
junction compensation are understood. Which is why, I
assume, the question does not specify the type thermocouple.
B0nz0 is correct; any type thermocouple sitting on a desk
at 25 deg C; hot end = 25 Deg C, cold end = 25 Deg C, will
read 0.0mV because there is no thermal gradient.
The cold junction temperature is the same as the hot end
temperature. Doesn't matter if it's a Type K or J or S.
With the same temperature at both ends, the output is 0mV.
But without specifying what the cold end temperature is, an
answer of 0mV assumes that the cold end is at 25 Deg C.
In the real world, it usually isn't.
The hot end can be at 25 Deg C with the cold end on a
terminal strip in control panel that's at 45 Deg C because
of the heat generated by the electronics in the control
panel. Then you'll read a negative voltage because the
cold end is hotter than the hot end!
The hot end can be at 25 Deg C and the cold end is
connected to handheld temperature calibrator in northern
Canada in the winter where the cold end terminal connection
might be at -30 Deg C.
In each case, it's the cold junction compensation that
corrects the temperature reading by measuring the
temperature of the cold end connection and taking that
temperature into account.
If I were to ask the question, I'd ask, "What is the output
of a thermocouple, both ends of which are at 25 deg C, and
why?". The answer to the first part is 0mV.
Tell 'em Carl sent you.
| Is This Answer Correct ? | 9 Yes | 0 No |
Answer / mohammed abbas
Itis depend upon the sensitivity of the thermocouple used
because the different combination of alloy is having
diffrent sensitivity like.
1,copper/contantan =0.039mv/deg
25deg c=0.915
2,Iron/contantan=0.52mv/deg
25deg c=13mv
3,Platinum/rhodium=.05mv/deg
25deg c=1.25mv
| Is This Answer Correct ? | 13 Yes | 5 No |
Answer / mohamed yousuf
it is depend upon the sensitivity of the thermocouple used
| Is This Answer Correct ? | 16 Yes | 13 No |
Answer / shabbir ahmad
first of all see it type it depends on the type of
thermocouple for example s type thermocouple has 0.113mv at
20degc every type of thermocouple has different output at
25degc.
| Is This Answer Correct ? | 1 Yes | 2 No |
Hello-I know this not an engineering question,I am working in gulf & i want to do correspondence diploma in instrumentation,can any one give me the address of any university in which this courses are available.
what will happen if transmitter wire get shoted?
What is scale factor of proximitor & unit of the viabration.
what is corrective maintenance , explain with difination
1 Answers Bapco Bahrain Petroleum Company,
What does a fire detector detects?
If a control valve is hunting means means if we are giving 50% to the cv, the cv will open at 51% then slowly slwoly it will reaches 49% , then it will reaches again 50%, this type of action occured frequntly.. so what to do for solve this problem. The possible cause is in volume booster or positioner's relay.. positioner is double acting .. and make siemens..sipart.The cv is double acting and mounted for anti surge operation, it having 4 nos. volume booster, one sov, one positioner, etc..?
2 Answers Gujarat Fluorochemicals Limited (GFL), Shell,
What is 3 selection available when configure the burnout protection for controller
0 Answers EI Electrotechnical Institute,
Witch type cable use for rtd ,why?
What is 1:1 redundancy ?
What is voltage free(wet) and potential (dry) contact?
why we use 24v dc in instrument rahter than 230vac????
what is PID n give me brief info. about P&I&D.
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)