How a pH sensor & analyzer work at waste water treatment
plant, whats the principle & science is beyond this,How pH
sensor sence & convert signal & move toward analyzer &
analyzer show reading in percent? also what is temperature
compensate option in analyzer
Answer / rajesh
pH is a measure of the acidity or basicity of a solution. pH
is a potentiometric measurement where the electrode used
produces an electrical signal which the pH meter converts to
a pH reading. The signal produced and measured is a voltage
and since voltage is a potential difference, two potentials
are needed to make a measurement:
a) The sensing electrode provides a potential proportional
to the logarithm of the hydrogen ion activity in the sample.
b) The reference electrode ideally provides a stable and
consistent potential independent of the activity of the sample.
The potential difference between the reference and sensing
electrode is measured by the meter and converted into a pH
value.
All pH sensors consist of a measuring glass electrode and
reference electrode. A combination electrode is an electrode
which contains both the reference and measuring electrode in
the one body. The reference electrode provides a constant
potential regardless of the composition of the solution it
is placed in. The voltage difference between the reference
and measuring electrodes is a function of the pH value of a
solution.
Calibration is done using pH Buffers. Most modern pH meters
are supplied with the facility for automatic buffer
recognition. The manufacturer will have programmed in a pH
versus temperature profile into memory for a specific type
of buffer. If a different buffer to that specified by the
manufacturer is used for calibration this will result in
calibration errors. For proper calibration, buffer solutions
of carefully selected accuracy and pH range are required.
The calibration should be performed using two buffers whose
values bracket the expected value of the samples being
measured. In practice many pH meters are equipped with auto
buffer recognition - this requires the use of the buffer
values specified by the meter manufacturer.
| Is This Answer Correct ? | 15 Yes | 1 No |
Paper Details: company -- Hindustan Copper Limited post -- GET 2008 Date -- 03.08.2008 Dept.-- Electronics?Instrumentation Aptitude -- 70 Technical-- 80 Total Time -- 2hr 30min I have listed the question I could able to remember 1. BCD equivalent of 43 2. value of K for stability of a system from a given equation 3. transformer connection for darlington connection 4. damping of balastic galvanometer 5. advantage of chart-pen recorder over CRo 6. CRO - how to change bandwidth 7. questions on CRO construction 8. reverse saturation current of si diode 9. advantage of wire wound resistance over carbon resistance 10. core material for a small coil 11. max fanout-- TTL/ECL/CMOS/MOS 12. component of resistance of monolithic IC 13. functions of capacitors in multivibrators 14. to change a sawtooth signal into a square wave we use-- schmit trigger/ schottkey diode/ zener diode 15. relation b/w á & â in transistor ckt 16. function of SiO2 layer in fabrication 17. 3 questions on ODD MAN OUT
I have energy meter for educational project having 0.5s class but my friend need 0.2s class meter , which one is good for that purpose and what is the actual difference of that two classes , need very easy example also.??
Can anybody tell me which type of question are asked in the interview for the FORMULATION(Pharma) as a instrumentation engineer, & question in building managment system.
0 Answers Reddy Labs, ScienceWay,
pls tell me whethere flame proof HMI's are available or every HMI is manufactured by considering flame proof standard.
plc trouble shootings?
what are the difference between conventional and smart transmitter.
1.Give example of analog and digital signal 2. What is yhe advantage of 8086 than 8085
5 Answers Bhel, College School Exams Tests, Profitec,
If a normal magnetic type flow transmitter that can measure flow, temperature and show this parameters on field, is it possible by any means to get this (flow and temperature) parameters info. On dcs? And if yes then with HART, FO communication or any other communication system?
1 Answers Process Control Technologies,
What are the interlocks we have to check during the start up of LP/HP compressor?
what is corrective maintenance , explain with difination
1 Answers Bapco Bahrain Petroleum Company,
Explain , how timer based pneumatic control valve will work ?
What is the relationship between GAIN and PROPORTIONAL BAND.
1 Answers Kharafi National, Sabic,
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)