IF HPLC CALIBRATION FAILED THEN WHAT YOU DO
Answers were Sorted based on User's Feedback
Answer / natraj kumar
Example HPLC calibration : - Suppose one parameter is
failed .keep the status board for Instrument breakdown.And
raise the OOC (Out of calibration )document take the
approval from QC and QA Head. After rectification or change
the part of the Instrument, check the failed parameter of
calibration.
If it is Passed, then check randomly impact assesment of
the previous analysis during this calibration period.If
there is no impact then u can proceed for regular
usage.Total process to be complete in 30 days only.
OOC document keep the calibration file. Service report is
also keep the filing internal or External service person.
Dr.Reddys Lab
Is This Answer Correct ? | 72 Yes | 3 No |
Answer / k.g.gupta dr.reddys
Give the Instrument for breakdown. After rectification of
the Instrument, check the whole parameters of calibrations.
If it is Passed, then check randomly impact assesment of
the previous analysis during this calibration period.If
there is no impact then u can proceed for regular usage.
Is This Answer Correct ? | 22 Yes | 5 No |
Answer / ajinkya p.
If hplc calibration fails, first we can cross check that all parameters are correct, sample preparation correct or not, system washed proper or not, condition of column is ok or not. first we will try to solve issue ourself. but if still calibration fails then discuss issue with senior person so that if there minor problem so that we can solve it ourself by their experience. but if any major problem we can take help by customer online support calling of instrument company person. bt still calibration fails we need to contact instrument engineer for visit.
Is This Answer Correct ? | 11 Yes | 0 No |
Answer / kuchelarao
ONCE U CAN CHECK THE PREVIOUS ANALYSIS AND IT IS IMPURITY
PROFILE AND THAT BATCH MUST BE COMMERCIAL AND ANALYSED TAHT
SAMPLE U CAN CHECK IT OUT PREVIOUS IMPURITY PROFILE BOTH
PROFILES ARE SAME THAT IS NO PROBLEM NOT MATCHED CAL THE
SERVICE ENGENEER.INFORM THE CUSTOMER......
Is This Answer Correct ? | 10 Yes | 12 No |
Principle of single pan analytical balance
0 Answers Amneal Pharmaceuticals, Bayer, Dr Reddys,
What r the limitatoins of uv visible spectroscopy ? Why Water use in blank standard for uv visible spectroscopy ?
Why we have to take 42gm of sodium hydroxide instead of 40gm for preparation of 1M Sodium hydroxide as per Pharmacopia? Whether it means the formula for calculating molarity was wrong?
5 Answers Claris Otsuka, Ranbaxy,
for which product require to do content uniformity? what is limit of cu?
why reference flow required during hplc analysis on rid?
what are the primary standards of redox titrations?Name any two.
how you fix the limits of impurities?
why we should take dst factor for below 1%moisture samples
0 Answers Ind Swift Laboratories,
How precise should be the linearity in HPLC method validation?
How you develop a method in HPLC?
0 Answers CU, Perrigo, Wockhardt,
Why sodium hydroxide used for maintain pH of phosphate buffer
what is the Bracketing & Matrixing.