Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...


IF HPLC CALIBRATION FAILED THEN WHAT YOU DO

Answers were Sorted based on User's Feedback



IF HPLC CALIBRATION FAILED THEN WHAT YOU DO..

Answer / natraj kumar

Example HPLC calibration : - Suppose one parameter is
failed .keep the status board for Instrument breakdown.And
raise the OOC (Out of calibration )document take the
approval from QC and QA Head. After rectification or change
the part of the Instrument, check the failed parameter of
calibration.
If it is Passed, then check randomly impact assesment of
the previous analysis during this calibration period.If
there is no impact then u can proceed for regular
usage.Total process to be complete in 30 days only.
OOC document keep the calibration file. Service report is
also keep the filing internal or External service person.

Dr.Reddys Lab

Is This Answer Correct ?    72 Yes 3 No

IF HPLC CALIBRATION FAILED THEN WHAT YOU DO..

Answer / k.g.gupta dr.reddys

Give the Instrument for breakdown. After rectification of
the Instrument, check the whole parameters of calibrations.
If it is Passed, then check randomly impact assesment of
the previous analysis during this calibration period.If
there is no impact then u can proceed for regular usage.

Is This Answer Correct ?    22 Yes 5 No

IF HPLC CALIBRATION FAILED THEN WHAT YOU DO..

Answer / ajinkya p.

If hplc calibration fails, first we can cross check that all parameters are correct, sample preparation correct or not, system washed proper or not, condition of column is ok or not. first we will try to solve issue ourself. but if still calibration fails then discuss issue with senior person so that if there minor problem so that we can solve it ourself by their experience. but if any major problem we can take help by customer online support calling of instrument company person. bt still calibration fails we need to contact instrument engineer for visit.

Is This Answer Correct ?    11 Yes 0 No

IF HPLC CALIBRATION FAILED THEN WHAT YOU DO..

Answer / u.venkatramireddy

call to system engineer

Is This Answer Correct ?    18 Yes 12 No

IF HPLC CALIBRATION FAILED THEN WHAT YOU DO..

Answer / suresh

calibration means whether the instrument properly working
or not, when compared to reference standard.

hplc calibration fail we can raise the quality impacting
insident that means which effect on the results of your
sample.

Is This Answer Correct ?    5 Yes 6 No

IF HPLC CALIBRATION FAILED THEN WHAT YOU DO..

Answer / kuchelarao

ONCE U CAN CHECK THE PREVIOUS ANALYSIS AND IT IS IMPURITY
PROFILE AND THAT BATCH MUST BE COMMERCIAL AND ANALYSED TAHT
SAMPLE U CAN CHECK IT OUT PREVIOUS IMPURITY PROFILE BOTH
PROFILES ARE SAME THAT IS NO PROBLEM NOT MATCHED CAL THE
SERVICE ENGENEER.INFORM THE CUSTOMER......

Is This Answer Correct ?    10 Yes 12 No

Post New Answer

More Analytical Chemistry Interview Questions

why equlirium conatant of water have 14 value

1 Answers   GVK Biosciences,


How much is the standard area for glc analysis

0 Answers  


what is range of Vivible UV ? What is UV& Vivible spctroscopy use for ?

6 Answers  


What is descriminatory disolution medium ?

3 Answers   Cipla, Graviti Pharmaceuticals,


how to select short coloum and long coloumn for new molecule

0 Answers   Rampex Labs,


How to fix the sample concentration in related substances development.

3 Answers   Novartis, Watson Pharmaceuticals,


on what basis SOP's can be changed? IF mobile phase ratio is 70:30 to how much % can we change the ratio in HPLC process

7 Answers   Cipla, Ranbaxy,


How are PDA Detector work in HPLC

7 Answers   Agilent,


what is SOP?How can prepare?

1 Answers   Reddy Labs,


difference between precision & accuracy.

1 Answers   Laurus,


what is the most common FT-IR source used for the MID&Near IR regions?

1 Answers   Dr Reddys, Piramal Enterprises,


why multimedia dissolution require to do?

0 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)