how to use K thermocouple instead E- type without
compesation cable
Answer / carl_ellis
Any thermocouple measuement requires 2 temperature measurements
- hot junction measurement
- cold junction measurement
A cold junction is created at the point where
non-compensation wire is connected to the thermocouple wire.
A measurement without compensation wire AND without a
corresponding cold junction measurement is not valid.
The only way to take a K type thermocouple measurement (or
any type thermocouple measurement) without compensating
cable is to use a temperature transmitter configured for the
type of thermocouple because the transmitter will perform
the cold junction measurement at its wiring terminals.
Carl Ellis
Measure First
Is This Answer Correct ? | 0 Yes | 0 No |
a pressure transmitter with a range of 0 to 2500kpag has a specified accuracy of 0.5 % of span.It is to be calibrated using a comparison gauge to have an accuracy of 0.1% of span.The comparison gauge has a range of 0 to 3000kpag What is the accuracy ratio? is this ratio acceptable?
How much should be the values and/or percentages of these given below factors for a good controller ? Steady state error Delay Rise Time Overshoot Peak Time Settling Time Damping Oscillations & vibrations thanks
What are the difference between mass flow meter, magnetic flow meter and vortex flow meter.
What is zero elevation and zero suppression?
Can a Differential Pressure Transmitter be used as a normal pressure transmitter??
How does pyrometer works?
WHAT IS THE DIFF BETWEEN FREQUENCY AND INTENSITY?HOW IT WORKS IN FLAME SCANNER ?
what is difference between wheatstone bridge, pirani gauge and strain gauge ??? explain its principle and working
Dear Friends, we are planning to design Power distribution and Fiber optic in the same cabinet.Like Front side Fiber optic and Rear side Power distribution.Is it any possibility of problem occurs like interference etc? Your answer is highly appreciate.
Close loop and open loop diagram for instrumentation
In vortex flow meter how can measure the frequency of vortex generated by vortex flow meter
is here anybody who knows how to calibrate the pressure transmitter? ONLINE