Answer / haresh kidecha
generally volatile buffer used in LC/MS mobile phase.buffer
used in mobile phase is ammonium acetate ,ammonium formate
in concentration of 2-10mM for positive ions mode and 2-
50mM for negative ion mode.
Is This Answer Correct ? | 14 Yes | 1 No |
Answer / chandra sekhar
In LCMS WE R USING ONLY VOLTILE BUFFER&CONC.OF BUFFER NOT
MORE THAN 0.05M&0.2%
Is This Answer Correct ? | 10 Yes | 1 No |
Answer / balaji
In lc-ms volitile buffers only use conc upto 10mM maxium
Is This Answer Correct ? | 3 Yes | 2 No |