Answer / richard marks
ph meters are calibrated using buffer solutins which are at
known ph levels. Make sure the probe is cleaned before
imerging it in the solutions.
Is This Answer Correct ? | 16 Yes | 2 No |
Answer / satish
YOU CAN CALIBRATE IN THREE PHASE 4/7/9 PH BUFFER SOLUTION
NORMALLY WE ARE CALIBRATING WITH 4 & 9 PH BUFFER THAT MEAN
4PH--4MA/9PH--20
Is This Answer Correct ? | 11 Yes | 2 No |
Answer / subhash
the Ph calibration is done by using the buffer solutions
,with the one probe i,e the reference electrode and the
measurement electrode ,,,,,,,,,,,,,
Is This Answer Correct ? | 4 Yes | 0 No |
How to calculate 47Liter H2 Gas cyilnder consumtion? Cylinder inlet Pressure is 200bar and Discharge pressure is 6.0bar. Flow is 100ml/min.For the above condition how many days one cylinder can be used?
Hot calibrate a level transmitter (wet & dry legs)?
what is the characteristics of control valve?
if we have 380 KVA DG. how to calculate Power Factor. With detailed formula
In dcs contol valve parameters showing like this Gain=1 or 2 , ti=7 sec, td=20sec. But temp,pressure,level control valves we are using in which type parameter side using ti,td clearly tell me
How we calculate orifice plat size???
how to calculate signal cable sizing? and multipair cable sizing??explain in brief?
hi all plz tell me difrence between analog or digital load cell.
what is lead lag compensator and for wat v using that???
What is the difference between conductive lined pipe and conductive unlined pipe? Why do we see that?
0 Answers Process Control Technologies,
Emergency Shutdown Systems/ Plant Shutdown System.
My rtd working well at the same time r/i converter is not working properly.when power is given to rtd..it is giving 30mA and getting 2.4 mA if i remove the compensate leads No matter how many ohms.why it is happening like so?what is the solution?rtd range-0-200deg c& 4-20 mA o/p.please give me solutiin anyone......