Rtd sensor how to calibrate
Answers were Sorted based on User's Feedback
Answer / ahmallam
RTD Sensor can't be calibrate , you can only calibrate the transmitter only by potentiometer according to PT100, PT1000, ...... table.
or you can use comparison calibration.
Is This Answer Correct ? | 10 Yes | 1 No |
Answer / carl_ellis
An RTD is 'calibrated' by heating or cooling it to various
temperatures, measuring its resistance at those
temperatures, and the recording and analyzing the measurment
results.
An RTD's performance can not be adjusted like a temperature
transmitter's output can be adjusted, but its performance
can indeed be measured in a periodic calibration.
Carl Ellis
Measure First
Is This Answer Correct ? | 7 Yes | 2 No |
Answer / ajay kumar gupta
An RTD is 'calibrated' with the help of Temp. bath by selecting required temprature. You can compare RTD resistance, RTD Chart accordind to type RTD. There is no adjustment in RTD.
Before installation RTD in field this checking is must.
Is This Answer Correct ? | 2 Yes | 0 No |
Answer / yugesh shah
Calibration of rtd is not possible , yeah just know about error is possible when other rtd same type using in same place ..
Is This Answer Correct ? | 0 Yes | 0 No |
Where we r using split range and ratio controller?
what is solenoid valve?
what is the daily job of an instrument commissioning supervisor
how the actually positioner of control valve working? where we gave 3-15psi pressure to I/P converter where it is firstly applied ?- on I/P to positioner then what happen? is control valve operate on 3-15 psi or its bench pressure?
what is loop1 and loop2 in MTS LT?
Please anyone answer my question, what is the difference between single acting and double acting control valve, in what condition we better choose single acting rather than double acting
Turbidity analyzer working principle ?
How to Calibrate Flammable gas detector in field
What is the difference between Error and Offset.
How many interlocks at turbine? Plz explain briefly
0 Answers MRPL Mangalore Refinery and Petrochemicals Limited,
What is the diffirence between Earth and Neutral and What is the required normal voltage between earth and neutral
in the construction field why hook up s.s.tubing for transmitter using upto 6mtrs of closed loop, but y not open loop nomore restriction