can i calibrate DP transmitter online (while Tx in Service)
if yes ,how?
Answer / carl_ellis
In order to 'wet'calibrate a transmitter (by applying
reference pressure), the transmitter needs to be powered,
but if the transmitter's output signal is an input to a
control loop or alarm, then those functions need to be
disabled (loop put in manual, alarm disabled) because the
transmitter's output is not representative of the process
during calibration.
The transmitter has to be isolated from the process, so that
cal pressure can be applied. This done through isolation
valving; typically a 3 or 5 valve manifold for a DP transmitter.
So, is a transmitter 'in service' if it's powered, but its
signal is not representative of the process and is, in fact,
isolated from the process?
My answer is no, a transmitter being calibrated is not 'in
service'.
Carl Ellis
Measure First
Is This Answer Correct ? | 3 Yes | 0 No |
Why trim span ? How?
Exapain Deatails of the PID controlller
what is the difference between(Flame proof/Explosion proof)& Fire proof instruments.Both are same or different & if different in which way because the transmitter which have approved ATEX flame proof enclouser but it use PTPE O ring & PTFE Head Gasket,so its (fire safe/fire proof)????????
How d.p. Transmitter can be applied to close tank?
What are the two states of a controller
Explain the use of hair spring in the pressure gauge.
I want to know the questions asked in ONGC interview(after being selected in the written exam) for Instrumentation branch.Kindly help.
Calibration procedure for electro magnetic flow meter
What is the Working principle of Dozing valve?
what is Instrumentation?
why two probes are used in turbine flowmeters?
In which principle Woodward Governor 505 work in power plant? What are the input of Woodward governor 505