yokogawa Temperature transmitter showing Error 15 which is Reverse Cal Fail (confirmation calculation result is bad) in manual they mentioned reasion That the output goes to the value set by hardware failure mode jumper. How can solve this problem ? anybody can say solution
Answer / carl_ellis
In the manual (from which you cited) the user action for
error 15 is 'contact service personnel'.
You can try that, but I suspect that it means, for all
practical purposes, replace the transmitter. I don't know
if Yoko offers a spare 'puck' that can be replaced in order
to salvage the housing and display, or not. You'll have
to ask.
You might try cycling power to it to see if an electrical
spike on the loop wiring adversely affected the processor,
and cycling the power might re-initialize it, but the
likelihood of success is very small.
When a transmitter goes to fail-safe mode, it's usually the
last act of the last performance, the actors are clasping
hands, and taking a final bow before the show closes.
tell 'em Carl sent you
| Is This Answer Correct ? | 3 Yes | 0 No |
what is all types of analyzer in water production plant?
What is re-latch the manual trigger in solenoid valve
0 Answers Valvitalia Systems Division,
admittance level sensor meaning?
how can we tune a PID. If it is by zeiger nichols method..please explain briefly with help of an example.
how magentic flow tx. vortex flow tx & ultrasonic flow tx works and its prinicle?its callibtration prosedure?
What's the difference between cold loop test and hot loop test?
what is power factor?
why we us isolation barrier in hazardous area how it works to protect the area and please clarify its current flow.
how can we calibate ph meter. how can we calibrat conductivimeter how can we calibrat massflowmeter & it s working.
What is difference between flame retardant & flame resistance cable? Among which type of cable is used for ESD & Fire and Gas system application?
When doing a liquid seal on a differential flow transmitter , most of the time the sealing liquid will be heavier then the process fluid. For wich application will the sealing liquid be lighter than the process fluid if transmitter is mounted below the process line.
Hi buddies! my name is sachin and i am working as a electrical and instrumentation engineer in a paper plant. right now i am facing a problem and if someone could help me it will be an immense pleasure to me. i am using AC 800M to synchronise DCS500 drives on paper machine and rewinder.communication is on PROFIBUS and we are using NPBA-01 module to link profibus to DC drives.But these days two of these drives are showing a error named FIELDBUS COMM.TIMED OUT and the drives get tripped.I have changed the NPBA module and PROFIBUS cable.So if anyone can do please help me out with this problem.I'll be highly grateful.
Civil Engineering (5086)
Mechanical Engineering (4453)
Electrical Engineering (16638)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)