yokogawa Temperature transmitter showing Error 15 which is Reverse Cal Fail (confirmation calculation result is bad) in manual they mentioned reasion That the output goes to the value set by hardware failure mode jumper. How can solve this problem ? anybody can say solution
Answer / carl_ellis
In the manual (from which you cited) the user action for
error 15 is 'contact service personnel'.
You can try that, but I suspect that it means, for all
practical purposes, replace the transmitter. I don't know
if Yoko offers a spare 'puck' that can be replaced in order
to salvage the housing and display, or not. You'll have
to ask.
You might try cycling power to it to see if an electrical
spike on the loop wiring adversely affected the processor,
and cycling the power might re-initialize it, but the
likelihood of success is very small.
When a transmitter goes to fail-safe mode, it's usually the
last act of the last performance, the actors are clasping
hands, and taking a final bow before the show closes.
tell 'em Carl sent you
Is This Answer Correct ? | 3 Yes | 0 No |
Why DP level transmitter preferred for closed vessels and Radar level transmitter for open tanks??
What is pressure,What is temperature,What is flow,What is level
how to change 4 wire rtd transmitter into a 3 wire rtd transmitter? as our transmitter is working for 4 wire rtd, what should do for working on 3 wire rtd transmitter
Why three wire is using with a RTD?
How to do ma trim in massflow meter?
what is gas metering system? gas chromatograph? and gas compressors?
During valve sizing from all calculation you got flow coefficient CV 12,than what it means? like for selecting pressure transmitter if your MAWP is 12 bar than you select a capsule of max. 20 bar which cover your mawp 12 in same way for selecting control valve If CV is 12 than how we use it in selecting valve??????
Why Differential pressure transmitter is located below the tappings points for steam & condensible vapor flow measurement application?
Why boiler chimney area oxyden sensor fixing?
What is the difference between HMI and SCADA?
how to calibrate the temprature transmitter MURCREYtype and thermocouple input type.
what is the relation between flow and pressure.for example relationship between level and pressure is p = phg like that i want ans for this question.and also how to calculate orifice sizing and flow transmitter. pls some one ans for this questions.