WHAT IS PVT TEST IN DISSOLUTION TEST , WHAT IS PVT SIGNIFICANCE AND HOW IT IS PERFORMED ?
Answer / archana
PVT Test in dissolution test means performance verification test which is also known as apparatus suitability test
Frequency of this test is once in six months.
This can be done by using prednisone USP Tablet 10 mg /tablet
for disintegrating type of tablets.
Current lot for this tablet is O0C056.
For Type 1 Apparatus
Dissolution medium :900 ml water
Temp :37°C
RPM :50
%dissolved :53-77%(O0C056)
Time :30mins
For Type 2 Apparatus
Dissolution medium :900ml water
Temp :37°C
RPM :50
%dissolved :27-48%(O0C056)
Time :30mins
Readings r taken at 242 nm
Std concentration 20 mg /ml of prednisone USP
Salicylic Acid Tablets are used for non disintegrating type of tablets
Current lot is Q0D200
For Type 1 Apparatus
Dissolution medium :900 ml of phosphate buffer of pH 7.4+/-.005
Temp :37°C
RPM :50
%dissolved :23-29%(Q0D200)
Time :30mins
For Type 1 Apparatus
Dissolution medium :900 ml of phosphate buffer of pH 7.4+/-.005
Temp :37°C
RPM :50
%dissolved :17-26%(Q0D200)
Time :30mins
Readings r taken at 296 nm
Std concentration 75 mg /ml of Salicylic acid USP
| Is This Answer Correct ? | 2 Yes | 0 No |
what is the importance of peak purity in HPLC and how we can calculate through manul(not software)?
What is control room temperature and which guide line says?
what is the defination of calibration & verification?
what is the meaning of ventillation in HPLC?
3 Answers Amneal Pharmaceuticals,
What is difference between Primary and secodary reference standard?
How can we calibrate the burette dispencing capacity in metrorm kf titration calibration
in dissolution in one tablets two molecule one is losartan & second is HCTZ losartam is complies in s1 stage & hctz is complies in s2 stage then how to report result losartan s1 stage result or losartan s2 stage result
Why we perform the ir analysis?
If change in specification which parameter require to do for validation? if change in chromtographic condition then which parameter? if api change then which parameter? if change composition then which parameter? if old method not work out then whicj parameter? if additional one impurity added then which parameter of validation require to do on above each conditions? elaborate separately
what is RI detector Cut-off Value
What is linearity in hplc and how to calibration
Why Peak Purity is not determined in GC analysis?