testing of thermo-couple?
Answers were Sorted based on User's Feedback
Answer / mithilesh kumar
For testing of any new Thermo couple in room temp., it's
checked by DMM.
1)DMM in contunity mode.
2) two wire is shows contunity.
for testing of thermocouple in high temp.
then calculations are follws these formula:
Degc= output of mv +ambient temp
where
1 mv= 26 degc..which shows in digital multimeter
if any other comment you most welcome.
good luck
| Is This Answer Correct ? | 1 Yes | 0 No |
Answer / sagar khandekar
Firstly you have checked 4 pair wire in the thermocouple,then after at adjacent pair checked by DC voltage in digital multimeter at room temperature,if they have show 2mV,it means thermocouple being ok for used.as chart prefer; 14C=1mV
| Is This Answer Correct ? | 2 Yes | 5 No |
Answer / sai kiran.c
actually thermocouple induces the emf .....which can be
tested by galvanometer.
actually emf induces abd thi scan be tested by dat
| Is This Answer Correct ? | 2 Yes | 9 No |
Why Instrumentation Air Compressor called as Instrumentation Air compressor?pls dont just tell its trasferring IA to all transmitter,valves.I need exact reason, please......................
Can you explain drawbacks between P,I,D & PID controller
1 Answers JSW, Priya Cement, TPL,
How can I read the electrical drawing easily?
What unit is used for measuring comp. outlet flow. it is CFM, M3/hr or NM3/hr. ?
What is the differences of Level swich( LS ) and Level Transmitter(LT)
Why we use 250 ohm resister in series hart communicator?
I want to know if flowrate increases in 16' pipe line what will happen to upstream pressure and why?
Name four general (not technical) criteria that would you consider important, when deciding upon a particular control valve?
How to calculate 47Liter H2 Gas cyilnder consumtion? Cylinder inlet Pressure is 200bar and Discharge pressure is 6.0bar. Flow is 100ml/min.For the above condition how many days one cylinder can be used?
Q1]IF THE RANGE FOR FLOW IS -35 TO 125 WHAT WILL BE THE SPAN FOR ABSOLUTE AND FOR NORMAL? Q2]HOW TO CALIBRATE CAPILLARY TYPE LT? Q3]HOW MANY ERROR CALCULATIONS ARE THERE USED IN PROCESS INDUSTRIES? Q4]HOW TO CALIBRATE TEMPERATURE TRANSMITTER? Q5] HOW TO CALIBRATE LEVEL TROLL AND HOW TO CONNECT? Q6]PLEASE TELL ME THE BASIC FOR INSTRUMENTATION WHICH IS APPLIED FOR CALIBRATION AND FOR ALL? PLEASE MAIL THE ANSWER TO honest_yes@yahoo.in
what is actual difference between pressure Transducer & Pressure Transmitter.
What is the basic difference between Modbus and CAN Open Protocols
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)