Answer / manikatna

UV-VIS spectrophotometer calibration is many parameters
which is
1)control of wavelength
2)control of absorbance
3)wavelength accuracy
4)straylight accuracy
5)resolution
6)resolution power

Is This Answer Correct ?    172 Yes 15 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / rakesh

1.0 CONTROL OF ABSORBANCES
PROCEDURE: REPARATION OF POTASSIUM DICROMATE SOLUTION: Dry
K2Cr2O7 at 130oC for 2 hours. Weigh accurately about the
quantity of sample equal to 60mg into 1000ml volumetric
flask. Add 0.005m Sulphuric acid to dissolve and dilute
with the same solvent.
Find out the absorbances of (A 1% 1cm) above said solution
at wavelengths of a 235nm, 257nm, 313nm and 350nm.
Calculate the (1% 1cm) value using the following expression.
Absorbance X 1 X 1000 / Weight of K2Cr2O7 in g X 100
Check the value for each wavelength with the tolerances
given four wavelengths.


PREPARATION OF REAGENT:
(i) Dilute 3ml of concentrated sulphuric acid and make up
100ml with distilled water.
(ii) Dilute 20ml to 2000ml with distilled water.
2. LIMIT OF STRAYLIGHT:
The absorbance of a 1.2-% w/v solution of potassium
chloride at a path-length of 1cm should be greater than 2.0
at about 200nm when compared with water as reference liquid.
1.2% POTASSIUM CHLORIDE SOLUTION:
Take 1.2 g of Potassium chloride previously dried at 105oC
for two hours into the 100ml volumetric flask. Add water to
dissolve and make up to volume with water.
3. RESOLUTION POWER:
Record the spectrum of a 0.02% v/v solution of Toluene in
Hexane UV. The ratio of the absorbance at the maximum at
about 269nm to that at the minimum at about 266nm is not
less than 1.5 unless other wise specified in the monograph.
0.02% v/v TOLUENE:
Take 0.02ml of Toluene into the 100ml volumetric flask. Add
Hexane to volume.

Is This Answer Correct ?    97 Yes 8 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / upk

UV-VIS spectrophotometer calibration is many parameters
which is
1)control of wavelength
2)control of absorbance
3)wavelength accuracy
4)straylight accuracy
5)resolution
6)resolution power &
7) Io Line Flatnes test

Is This Answer Correct ?    55 Yes 22 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / nitesh pal

in above answer also added 430 nm in control of wavelength.(acc to ipm dilution is about 60 mg to 1ooml (same solvent)

Is This Answer Correct ?    21 Yes 8 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / ramkrishna jagtap

control of wavelength

Is This Answer Correct ?    7 Yes 0 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / devender justa

Internal Calibration ,external Calibration, Baseline
Flatness

Is This Answer Correct ?    8 Yes 4 No

Post New Answer

More Analytical Chemistry Interview Questions

Why only hydroxy naphthol blue indicator is used for standardisation of 0.05M EDTA solution instead of solochrome Black T or Euriochrome Black T indicator which is used for all sample analysis with 0.05M EDTA solution?

0 Answers  


why phenolpthalin gives pink colour when we titrate aganist sodium hydroxcide & phenolpthalin gives colourless titrate aganist hydrochlric acid which is un burate

1 Answers  


what is mean by covalidation

0 Answers  


In which cases factor of KF Reaget got higher side

2 Answers  


why we use only vanilline , acetanilide,& caffeine For the Calibration of melting point appartus?

7 Answers   Reddy Labs,






why ph is between 0 to 14

6 Answers   Apotex, Reliance, Zydus Cadila,


What is related substance by HPLC impurity limits as per USP?

0 Answers  


WHAT IS DIFFERENCE BETWEEN RRT AND RRF IN HPLC

2 Answers   Cadila,


Why are use silicon oil mr?

0 Answers  


why we use actone for Gpv test in hplc calibration

1 Answers  


why we use 0.1% acetone in gradient calibration of HPLC?

3 Answers   Cipla, Dr Reddys, FDC,


define beer-lambertslaw.

5 Answers   Laurus,


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)