1.0 CONTROL OF ABSORBANCES
PROCEDURE: REPARATION OF POTASSIUM DICROMATE SOLUTION: Dry
K2Cr2O7 at 130oC for 2 hours. Weigh accurately about the
quantity of sample equal to 60mg into 1000ml volumetric
flask. Add 0.005m Sulphuric acid to dissolve and dilute
with the same solvent.
Find out the absorbances of (A 1% 1cm) above said solution
at wavelengths of a 235nm, 257nm, 313nm and 350nm.
Calculate the (1% 1cm) value using the following expression.
Absorbance X 1 X 1000 / Weight of K2Cr2O7 in g X 100
Check the value for each wavelength with the tolerances
given four wavelengths.
PREPARATION OF REAGENT:
(i) Dilute 3ml of concentrated sulphuric acid and make up
100ml with distilled water.
(ii) Dilute 20ml to 2000ml with distilled water.
2. LIMIT OF STRAYLIGHT:
The absorbance of a 1.2-% w/v solution of potassium
chloride at a path-length of 1cm should be greater than 2.0
at about 200nm when compared with water as reference liquid.
1.2% POTASSIUM CHLORIDE SOLUTION:
Take 1.2 g of Potassium chloride previously dried at 105oC
for two hours into the 100ml volumetric flask. Add water to
dissolve and make up to volume with water.
3. RESOLUTION POWER:
Record the spectrum of a 0.02% v/v solution of Toluene in
Hexane UV. The ratio of the absorbance at the maximum at
about 269nm to that at the minimum at about 266nm is not
less than 1.5 unless other wise specified in the monograph.
0.02% v/v TOLUENE:
Take 0.02ml of Toluene into the 100ml volumetric flask. Add
Hexane to volume.
Is This Answer Correct ? | 97 Yes | 8 No |
Answer / upk
UV-VIS spectrophotometer calibration is many parameters
which is
1)control of wavelength
2)control of absorbance
3)wavelength accuracy
4)straylight accuracy
5)resolution
6)resolution power &
7) Io Line Flatnes test
Is This Answer Correct ? | 55 Yes | 22 No |
Answer / nitesh pal
in above answer also added 430 nm in control of wavelength.(acc to ipm dilution is about 60 mg to 1ooml (same solvent)
Is This Answer Correct ? | 21 Yes | 8 No |
Answer / devender justa
Internal Calibration ,external Calibration, Baseline
Flatness
Is This Answer Correct ? | 8 Yes | 4 No |
Why only hydroxy naphthol blue indicator is used for standardisation of 0.05M EDTA solution instead of solochrome Black T or Euriochrome Black T indicator which is used for all sample analysis with 0.05M EDTA solution?
why phenolpthalin gives pink colour when we titrate aganist sodium hydroxcide & phenolpthalin gives colourless titrate aganist hydrochlric acid which is un burate
what is mean by covalidation
In which cases factor of KF Reaget got higher side
why we use only vanilline , acetanilide,& caffeine For the Calibration of melting point appartus?
why ph is between 0 to 14
6 Answers Apotex, Reliance, Zydus Cadila,
What is related substance by HPLC impurity limits as per USP?
WHAT IS DIFFERENCE BETWEEN RRT AND RRF IN HPLC
Why are use silicon oil mr?
why we use actone for Gpv test in hplc calibration
why we use 0.1% acetone in gradient calibration of HPLC?
3 Answers Cipla, Dr Reddys, FDC,
define beer-lambertslaw.