Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...


What is the procedure for UV-spectrophotometer calibration?

Answers were Sorted based on User's Feedback



What is the procedure for UV-spectrophotometer calibration?..

Answer / manikatna

UV-VIS spectrophotometer calibration is many parameters
which is
1)control of wavelength
2)control of absorbance
3)wavelength accuracy
4)straylight accuracy
5)resolution
6)resolution power

Is This Answer Correct ?    172 Yes 15 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / rakesh

1.0 CONTROL OF ABSORBANCES
PROCEDURE: REPARATION OF POTASSIUM DICROMATE SOLUTION: Dry
K2Cr2O7 at 130oC for 2 hours. Weigh accurately about the
quantity of sample equal to 60mg into 1000ml volumetric
flask. Add 0.005m Sulphuric acid to dissolve and dilute
with the same solvent.
Find out the absorbances of (A 1% 1cm) above said solution
at wavelengths of a 235nm, 257nm, 313nm and 350nm.
Calculate the (1% 1cm) value using the following expression.
Absorbance X 1 X 1000 / Weight of K2Cr2O7 in g X 100
Check the value for each wavelength with the tolerances
given four wavelengths.


PREPARATION OF REAGENT:
(i) Dilute 3ml of concentrated sulphuric acid and make up
100ml with distilled water.
(ii) Dilute 20ml to 2000ml with distilled water.
2. LIMIT OF STRAYLIGHT:
The absorbance of a 1.2-% w/v solution of potassium
chloride at a path-length of 1cm should be greater than 2.0
at about 200nm when compared with water as reference liquid.
1.2% POTASSIUM CHLORIDE SOLUTION:
Take 1.2 g of Potassium chloride previously dried at 105oC
for two hours into the 100ml volumetric flask. Add water to
dissolve and make up to volume with water.
3. RESOLUTION POWER:
Record the spectrum of a 0.02% v/v solution of Toluene in
Hexane UV. The ratio of the absorbance at the maximum at
about 269nm to that at the minimum at about 266nm is not
less than 1.5 unless other wise specified in the monograph.
0.02% v/v TOLUENE:
Take 0.02ml of Toluene into the 100ml volumetric flask. Add
Hexane to volume.

Is This Answer Correct ?    97 Yes 8 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / upk

UV-VIS spectrophotometer calibration is many parameters
which is
1)control of wavelength
2)control of absorbance
3)wavelength accuracy
4)straylight accuracy
5)resolution
6)resolution power &
7) Io Line Flatnes test

Is This Answer Correct ?    55 Yes 22 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / nitesh pal

in above answer also added 430 nm in control of wavelength.(acc to ipm dilution is about 60 mg to 1ooml (same solvent)

Is This Answer Correct ?    21 Yes 8 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / ramkrishna jagtap

control of wavelength

Is This Answer Correct ?    7 Yes 0 No

What is the procedure for UV-spectrophotometer calibration?..

Answer / devender justa

Internal Calibration ,external Calibration, Baseline
Flatness

Is This Answer Correct ?    8 Yes 4 No

Post New Answer

More Analytical Chemistry Interview Questions

Why Deuterium lamp is used as UV light Source for UV.

4 Answers  


What is primary & Secondary Standard ? Give Example

17 Answers   Cipla, Lupin,


what is the basic principle of conductivity meter

4 Answers   Hetero,


how will you do the prep for unstable componds?

0 Answers   AMRI, CTX Life Sciences, Piramal Healthcare,


IS THERE ANY EQUIPMENT TO CHECK AND CALCULATE THE POLARITY OF A LIQUID?

0 Answers   Reddy Labs,


what king of work is done in Regulatory Affairs in Pharma industry ?

3 Answers   Quintiles,


what is dissolution What is unit of KF? what is the significant of pH in HPlc? WHY POTASSIUM HYDROGEN PTHALATE IS USED FOR PERCHLORIC ACID? What is difference btw UV detector & RID detector?

1 Answers  


Why the same type of columns showing different RT for a single product?

6 Answers  


why r u using hplc grade solvents

4 Answers  


why using Potentometry What Purpose? differance between manual and potentiometry? why using Some of materials in usig potentiometry? What is use of mericuric acetate in Potentiometry

1 Answers   Biocon,


what is the difference between BDS and ODS Columens and both are same or not

2 Answers   Intas, IPQA, Micro Labs, Vasava Engineering Private Limited,


How to calibration of the uv spectroscopy and its test?

0 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)