how to calibrate AAS ? give me the procedure?
Answers were Sorted based on User's Feedback
Answer / ch. appala naidu
AAS instrument is calibrate by Copper standard solution.
Usually 5 ppm copper standrd solution give absorbance 0.5
to 0.6 atomic units(a.u.)
The 5 ppm Copper standard solution should be prepare from
traceable level.
Is This Answer Correct ? | 12 Yes | 2 No |
Answer / ravi kumar kadari
5 ppm copper standard solution gives absorbance 0.5 atomic
units,And also measure the RSD,and RSD Should be less than
5 %
Is This Answer Correct ? | 5 Yes | 0 No |
Answer / nalini
we have gone through the AAS calibation procedure with copper,what is the reference method & no.
Is This Answer Correct ? | 0 Yes | 0 No |
why in GC calibration of linearity regression coefficient should not be less than 0.999
What is difference between Potency & purity
What is Log p?
how calibrate the hplc detector
2 Answers Cipla, Shraddha Analytical Services,
why base line not starts from zero in gc?
1 Answers Aurobindo, FDC, Reddy Labs,
why we are used 4,7,9,2 buffers for the ph calibration
4 Answers Aurobindo, Enaltec Labs,
what is the difference between potentiometric titration and karl fischer titration?
differnc btw c18 and c-8 column
is a bonded water molecule in an organic material can be easlisy and accuractly determined by KF rather than LOD
How to select buffer for HPLC mobile phase
4 Answers Cadila, Cipla, Lupin,
what is the d/f b/w normality and molarity and molality..?
What are the primary considerations in extractables and Leachables ?