Describe how to calibrate various types of instruments and
transmitters using ISA industrial consensus standard
techniques.
Answers were Sorted based on User's Feedback
Answer / amith.patil17@gmail.com
Calibration Can be Carried out By Calibration Instruments
such As For Pressure Transmitter,level transmitter the
pressure pumps with mA in Series or Can be used simply
HEART with a Pressure pump,By increasing the pump pressure
the master guage shows the indication of the Standared
pressure Corresponing to which mA reading canbe Obtained
check for incrasing and decreasing Valus for 4,8,12,16mA If
Error Occurs then Set a Zero or Span Adjustment,Simillarly
For RTD,T/C in Furnace is Carried out.
| Is This Answer Correct ? | 16 Yes | 1 No |
Answer / ram
Calibrate to any instruments like pressure transmitter by using the druk calibrator to build the pressure depends on transmitter range and measure the 4-20 ma it's increase or decrease have a span and zero settings .0_30bar like initially 4ma,8ma,12ma,16,20ma.and using hart communicator also
| Is This Answer Correct ? | 3 Yes | 0 No |
Calibration of instruments are carried out under master equipment,they are calibrated under the "National Instrumentation Laboratory".
I can not tell u about the detail because it is under rule But i will explain it with example:
To calibrate a temperature Gauge-
The master equipment use is "Fluidless temperature Calibrate".
it work With out fuel so it is called Fluidless Temperature Calibrator the only thing u have to do is Put the End nod into the master equipment and initially set the temp. to 0degree cent. and step up to 30degree cent. with this see the reading proportionally in the Gauge no need to calibrate if it give +-0.2 reading. if it give deviation of +-5 then adjust the span and zero to set point result we been seen these is the method of calibration of temperature gauge.
Other example of master equipment are microprocessor based pressure calibrator,Macloid gauge etc.
| Is This Answer Correct ? | 0 Yes | 3 No |
while am calibrating Drager polytron 2IR flame gas detector occured non zero status.how can i get zero status?
IF ASK IN THE INTENVIEW WHAT TYPE OF BMS YOU USE?
hi this is noor i wnt know the procedure and precations errection for PRESSURE AND DP transmitters? how to take impusle line for level and flow?(water and steam) impulse line sizes? and type of material ? plz give me answer / reffer any website ? to me "mail2noor.hyd@gmail.com"
Why thermo wells are used? What materials are used in thermo wells?
how many types of preheaters are in cement industry & what r they can any body explain?
how pressure reduce through pressure control valve? suppose we have to reduce 20 kg pressure to 10 kg pressure , then what we do ? open the valve or close the valve , initially it is 20 % is upon ( for ur knowledge ).
how manys ways dp transmitter can be calibrated having range of -150 to 400mmwc,what side pressure u will apply?
what are the sensor types of vibration&speed?n what is proximity sensor?
Can SmartPlant Instrumentation be used in a concurrent engineering environment?
Can anybody tell me application of PLC,DCS,SCADA.
pls suggest suitable MOC of diaphragm for pressure measurement of UREA grade solution service fluid.
what is the meaning of potential free contact. difference between dry and wet contact?
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)