Impossible to say, because the technology is not stated, nor are the conditions stated.
Yokogawa is a brand name, not a technology, and the reason for a negative flow indication depends on the technology.
Technologies for mass flow (however inferred) are
- multivariable flow meters based on pressure and temperature compensated differential pressure
- Coriolis flow meters
- Vortex flow meters with temperature and pressure compensation
- Thermal dispersion flow meters
which might or might not be Yokogawa products.
The reasons for a false negative flow indication could be quite different for any of the technologies
Conditions like the expected flow rate vs the observed flow rate, medium, medium temperature and pressure, service history, installation details (vertical or horizontal), flow path, type of pump, process function or contiguous piping and its orientation, and configuration elements like low flow cutoff could contribute to an indicated negative flow rate, or operational method of zeroing.
Because it could be that you actually have reverse flow from a thermal syphon or a pressure syphon.
Carl Ellis
Measure First
| Is This Answer Correct ? | 4 Yes | 0 No |
Answer / sarin
HP LP Ports will be interchanged. High pressure is coming in LP Side.
| Is This Answer Correct ? | 2 Yes | 2 No |
A differential transmitter is calibrated to measure the flow of lquied, explain what wil happen if the fluid is not pure but contains some bubble?
typical loop drawing of a simple feedback control comprising of one transmitter and one control
1. what is 3 point calibration & 5 point calibration? 2. If we use hart communicator to set the LRV & URV of a smart transmitter, did we calibrate the transmitter? why? 3. Make a complet list if meterials needed to properly connect a pressure gauge to a steam pipe using 1/2 inch stainless steel tubing. The gauge connections is 1/2 inch NPT-M and the steam pipe tap has a 1/2 inch female coupling, compression fittings are use. 4. What are the tools in loop checking? 5. What would the ISA symbol of a square root extractor look like? 6. What would you do to completely prepare for the configuration? 7. What safety measures must you undertake when doing configuration? 8. what are the two most commonly use programming languages used when we configure a PLC? 9. What would you do if you accidentally skipped a step in the configuration procedure? 10. How would you know if the configuration were sucessful? 11. Name at least 3 selections available when we configure the burnout protection of a controller?
3 Answers EI Electrotechnical Institute,
Name some of the common protocol use in automation.
how can i calibrate the temperatuture gauge used by temperature bath(dry block) i mean step by step procedure?
What is the standard height for stancion supports? and what is the standard mounting height for different transmitters?
what is see back effect?
what is load line in control valve?
3 Answers Boral, Qatar Petroleums,
what are the possible reasons if a control valve fails to open/close?
difference b\w a transducer and a sensor
Wht is Dry Leg Calibration and Wet Leg Calibration?
why we use impulse line of flow transmitter at 45 deg. angle both side of orifice plate.
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)