How to Calibrate Transducers? Can you give me a step by ste
procedure?
Answer / mahesh bajaj
The transducers are calibrated against reference standards
which are atleast 10 times more accurate then the transducer
to be calibrated.
Let me elaborate by giving an example.
Orifice plate infers flow rate by measuring differential
pressure.so to calibrate orifice plate, reference standard
say magnetic flowmeter is used which measures flow rate.say
flow range is 0-100 lpm and at every 10 lpm, differential
pressure across orifice is measured and constant is evolved
between flow rate and differential pressure. This constant
is also known as discharge coefficient or calibration
constant and one dimensional bernolli's equation is used to
compute this constant.
Now once you know the constant the orifice flow element is
ready to be used in any line size without using any
reference standard.
Hope this helps.
Is This Answer Correct ? | 7 Yes | 1 No |
hi my name is rakesh patel i just complited my be in instrumentation & control now i am going to working as a validation engineer in a poject department so what is a scope in a future as a validation engginer ? what line is good for me can any bady give me complete solution please? your regard rakesh patel
What is furnace draft control?
how many types of preheaters are in cement industry & what r they can any body explain?
which type of zones used for boiler? & why ?
how to know power plant process with corresbonding laws&thoeries
how u calculate ai/di/ao/do capacity in ur dcs?
I am new in work for 1 of our project we have some instrument in BOM including Pressure, Temp. and so on, I also noticed some objects like pressure switch process HOOK UP; which I don't know what does "HOOK UP" mean here.
What is Droop? What is Step/Gradient Limiter in Turbine Governor control?
In field side transmitter voltage droping here no loose connection then what will happen?
While preparing the graphic diagrams trend display, 3 types of trends are prepared by Yokogawa system vender. These are for process variable, manipulated variable & set point variable. Why these 3 trends are required? Explain the purpose of these 3 trends.
What is pressure,What is temperature,What is flow,What is level
Can anybody explain what is the different between ESD & PLC