Answer / n.srikanth
The Spectral Bandwidth (SBW) of a spectrophotometer is the
basis of establishing its ability to resolve absorbtion
lines seperated by small differences in wavelength. If a
substance that you need to measure has an absorbtion line at
257.4nm and there is an interfering line at 260.0 nm you
need to establish whether your spectrophotometer can resolve
the two lines seperately or "mixes" the two absorbtion lines
into a single line which would give erroneous results.
Toluene in Hexane is used as a reference for the calculation
of SBW. The ratio of the absorbtion of the solution when
read at a maxima (around 268.7nm) and at a minima (around
267.0nm) relates directly to the SBW of the instrument being
assessed. Regular use of this technique will assure that the
resolution of your instrument is within the required range
for your work
| Is This Answer Correct ? | 4 Yes | 2 No |
related substances test replicate injection is allow or not if allow so rsd limit how many %?
What is GMP, GLP?????????
diffrence between known impurity and specified impurity?
How to preform dissolution profiling
What skills have you acquired that would apply to the position of Clinical Data Manager?
what is the diffirence between deviation and change control? at what situations we can raise the deviation and change control
31 Answers Biocon, Cipla, DRL, Glenmark, Granules, Lupin, Micro Labs, Mylan, Natco, Ranbaxy, Swissmedic,
If calibration of 12 bowl dissolution apparatus does not meets single stage procedure, how can u proceed calibration?
What is forced degradation study?
What is the differance between change deviation and change controle with respect to QA?
4 Answers Cipla, Ranbaxy, Wockhardt,
exolain th function of the joints
Tell us about a situation in which you took the extra step for a patient.
0 Answers Claris Lifesciences,
what are the principles of immunisation?