why use toline in resolution in uv calibration



why use toline in resolution in uv calibration..

Answer / n.srikanth

The Spectral Bandwidth (SBW) of a spectrophotometer is the
basis of establishing its ability to resolve absorbtion
lines seperated by small differences in wavelength. If a
substance that you need to measure has an absorbtion line at
257.4nm and there is an interfering line at 260.0 nm you
need to establish whether your spectrophotometer can resolve
the two lines seperately or "mixes" the two absorbtion lines
into a single line which would give erroneous results.

Toluene in Hexane is used as a reference for the calculation
of SBW. The ratio of the absorbtion of the solution when
read at a maxima (around 268.7nm) and at a minima (around
267.0nm) relates directly to the SBW of the instrument being
assessed. Regular use of this technique will assure that the
resolution of your instrument is within the required range
for your work

Is This Answer Correct ?    4 Yes 2 No

Post New Answer

More Health Pharma AllOther Interview Questions

plz.mail me model papers of drug inspector of chhattisgarh..or MP.if Posible. Mail Id-recha_1428@rediffmail.com...mrs_1428@rediffmail.com,

0 Answers  


What is forced degradation study?

0 Answers  


how many clauses are there in iso? give me the clauses details?

0 Answers  


Caffeine standard is recommended by the monograph for HPLC wavelength accuracy calibration. It is covered range of 205 to 273 only.If the method have wavelength beyond the range, then how it supports for instrument performance?

0 Answers  


What is the role of Pharma QA and Validation Tester - Referring to jobs as Computer system validation analyst - interview quetions.

2 Answers  






pls send me last two years solved papers of health & malaria inspector to kumar.santosh087@yahoo.com

0 Answers   IPCA, RRB,


What was your most difficult experience with this population and how did you deal with it?

0 Answers  


What is the Difference between Humidity and Relative Humidity ?

2 Answers   Sanmar,


Give an example of how you've handled a problem patient.

0 Answers  


when we connect the capacitor banks in series with the circuit? 2 144 pH is having unit?

0 Answers  


human body absorbs maximum amount of nutrients in

1 Answers  


why u want join cipla company as a medical representative

3 Answers   Alembic, Cipla,


Categories