What is use of LOD and LOQ in validtion and give range and in calculation why we use 3.3* LOD here why we use 3.3
Answer Posted / nikkosthiru@gmail.corai12
LOD is Limit of detection. We can dedect the peak but cant quantify
LOQ Limit od quaantification. We quantify also this is known as reporting thereshold.
During the RS analysis Below LOQ level we cant included in the total impurities calculation.
LOD and LOQ Determination three methods
1.S/N Ratio method
3 or 2 ratio for LOD
NLT 10 ratio for LOQ
2. Slope Method
Plot the linearity for active and impurities through serial dilution at lower level
Formula for LOD
3.3*Slope/Std.dev
Formula for LOQ
10*Slope/Std.Dev
3. Visualization method
It will be using for only identification and TLC Method
| Is This Answer Correct ? | 1 Yes | 0 No |
Post New Answer View All Answers
what you do when compound having dwel wavelength?
How to calculate coreletion coefficient
What is aggregate and fragments in SEC?
in which situation require to change rs specification?
From where require to take the RLD sample?
in dissolution if one bowl got 70 percent 2nd bowl got 80 percent and 3rd bowl got 90 percent then how proceed?
Tell me about analytical method validation in QC
what is mean by peak intigrity?
Please tell me about the pH of Polycaboxylic ether is it in the 5-6 range ever or more than 6
in gas chromatography what is the difference between gas flow rate and average linear velocity ?
In which situation we require to analytical method validation of excipient?
on saturation solubility study data how we can find out the bcs class of drug?
some product having water content method with pyridine and ethylene glycol mixture instead of methanol? how they are selecting and methanol is not suitable solvent how ?
identification is for unknown? qualification for known? reporting for LOQ?
How can we confirm the HPLC column is end-capped or not? Is it possible to identify by physical appearance?