How to prepare 10% HCl from 36.5% HCl solution.
Answer Posted / ron
The units % v/v or % wt/v need to be specified.
Many vendors sell 10% v/v HCl, which generally means 10 mL of concentrated (37%) HCl diluted to 100 mL of solution with water.
If it is 10% wt/vol desired based on true or absolute content of HCl then 10 g HCl x 1/36.5% = 27.4 g conc. HCl is added to a flask and diluted with water to 100 mL of solution.
Lets not worry about differences in the second decimal
place.
Sometimes molarity is less prone to confusion. Concentrated HCl is 12.2 molar. 10 mL of concentrated HCl diluted to 100 mL would give a 1.2 molar solution. There is no ambiguity in volume, mass, etc. with molarity.
Is This Answer Correct ? | 32 Yes | 2 No |
Post New Answer View All Answers
how can give the expiry period and restadardisation of volumetric solution
in which situation require to take incident in validation?
how will you do the prep for unstable componds?
How to establish relative response factor for hplc. Why it is required?
how you start RS method development when for new product?
What is related substance by HPLC impurity limits as per USP?
why we use glass fiber filters use in some situation?
Why to use propyl paraben in HPLC gradient calibration
Ratio of Water and Acetonitrile(diluent) is not recommended to use in cleaning method validation,why?
Which are the diffrent grades of api in pharma?
for which product require to do content uniformity? what is limit of cu?
why does used dry caffien in HPLC calibration?
What is the difference between spectro meter and spectro photo meter?
Why only Copper standard is used to calibrate Atomic Absorption spectrophotometer?
I have compare C2H2-air and C2H2-N2O flame AAS on determination calcium. Both use same range of std to plot calibration curve. (2-6ppm) When i measure the sample with phosphate, KCl and LaCl, C2H2-N2O flame give false positive result, around 0.5ppm. When i measure the sample with phosphste, KCl and EDTA. C2H2-N2O flame also give 0.5ppm false positive. But both above mentioned sample would not give false positive when measured by C2H2-air flame. What is the reason?