Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...

what is the difference between system precision and method
precision?

Answer Posted / ilayaraja

#1 System precision (System evaluation)

When you inject single preparation of standard in ten
times/or any number of replicates to the system and the
final result i.e %RSD meets acceptance criteria it shows
that, from injection , elution to detection of your
instrument functioning correctly.

Simply the system precision is evaluation of system.

#2 Method Precision (Method evaluation)


When you inject six preparation of same batch sample in
to your evaluated system,obtained six results should
precise.Then the method used for quantification is correct.

Simply the method precision is one of the testing parameter
to check the method used for the test is reproduble of
results or not.

Don't mitake if the answer is not up to the mark.

Is This Answer Correct ?    20 Yes 1 No



Post New Answer       View All Answers


Please Help Members By Posting Answers For Below Questions

how you confirm the assay method?

719


How can I develope method of dissolution by HPLC OR UV

2341


In HPLC (Waters) software Empower1,2 &3 what is the updation for each version

2954


How do we quantify crystaline and amarpous forms by using (NMR, XRD)spectroscopic techniques? Which any others instruments are useful for this quantification? explain

2338


I have compare C2H2-air and C2H2-N2O flame AAS on determination calcium. Both use same range of std to plot calibration curve. (2-6ppm) When i measure the sample with phosphate, KCl and LaCl, C2H2-N2O flame give false positive result, around 0.5ppm. When i measure the sample with phosphste, KCl and EDTA. C2H2-N2O flame also give 0.5ppm false positive. But both above mentioned sample would not give false positive when measured by C2H2-air flame. What is the reason?

3485


As per ICH related substances stability trend limit from initial to shelf life

3179


In GC calibration,what is the procedure to calibrate the TCD (thermal counductivity detector)suggest ?

3116


What is the difference between spectro meter and spectro photo meter?

11335


what is mean by covalidation

729


In BP2013, Loperamid HCl monograph. Assay by titration with 0.1N sodium hydroxide using hydrocloric acid 0.01N and reading the volume added between the 2 points of inflexion. I have a question that if the diluent solvent is ethanol is certainly consumed a amount of volume of titrant, so this volume must be eliminated on the result calculation or not apart from first point which is subtracted above.

3065


How do we get end points and how many end points are possible for citric acid and di-acid not theorotically answer should be given practically.

2537


If combination product how require to identify which imp is of which api?

786


how can I make the copper get white bye any salt or acit?

2146


What is the use of tlc and hplc? And when and where use?

1093


how we can identify the impurity is coming below loq at transfering site?

807