Why linearity of detector performed in HPLC calibration
Answer Posted / radheraman singh
Checking the detector response with know concentration of
of standard ,to ensure that your detector response is
directly related to your concentration.i.e response should
be linearly increase with concentration.
Is This Answer Correct ? | 51 Yes | 0 No |
Post New Answer View All Answers
how can I make the copper get white bye any salt or acit?
how you start RS method development when for new product?
what is mean by extactable and leachable study?
What is the formula to determine the concentration of M of a solution given the % transmittance? Use %T = 43.7 as an example.
what is difference between UV - VISIBLE MODEL NO like 1600,1601,1700 etc ? plz explain me
if content uniformity passing but dissolution varrying then what is next step?
How to set analyticl specification for combination products?
For standardization of volumetric solutions, acceptance limit for concentration is considered as 10% of nominal concentration and %RSD as NMT 0.2% (for appropriate no. of replicates). But , if %RSD has not met, what should be further action?
if instrument calibration fails then what require to do?
My question about gas chromatography sulfur chemiluminsecence detector. I test unknown sample gas by GC-SCD (calibrated ) and the result of *H2S is 279 PPM , *but when I test the same sample with the GC-TCD (calibrated ) the value of *H2S is 2500 PPM . I'd like to inform you that both GCs are calibrated and have very good operation conditions with stable parameters . the question is if the sample gas with higher H2S over detection limits of SCD detector (1000 ppm). why the result it 279 ppm Best regards
How they found 1mL of K.F reagent is equivalent to 5mg of water and if we change the composition of K.F reagent, is it can neutralize more amount of water?
Why only Copper standard is used to calibrate Atomic Absorption spectrophotometer?
How require to develop GC method? how to select diluent, GAS, column selection and other chromatographic conditions?
what type of question will ask in the interview of lab chemist.
calibrtion procedue of LC-MS