how FTIR calibration Doing?
Answers were Sorted based on User's Feedback
Answer / murugesan
with polystyrene film
wave number check and resolution as per BP
| Is This Answer Correct ? | 19 Yes | 1 No |
Answer / ramkrishna thakare
1)with polystyrene film
2)wave number check
3 ) resolution.
4) energy efficiency.
Ramkrishna thakare --------Sandoz
| Is This Answer Correct ? | 11 Yes | 2 No |
Answer / ramu
Doing the calibration of FT-IR using Polystyrene film.You
have to do the three things in the calibration procrdure.
1)Verification of Wave number scale
2)Control of resolution between two types you have to check
a)Transmission minimum at 1583
Transmission maximum at 1589
b)Transmission minimum at 2849.5
Transmission maximum at 2870
| Is This Answer Correct ? | 9 Yes | 1 No |
What is disintegration time for dispersible tablets?
What is main of k*abs in uv visible spectroscopy
Difference between Group-1 and Group-2 Exams
What types of columns are used in GC?
what is difference between usfda and mhra guidline
What is 21 CFR? And its significance.?
In resolution test of uv calibration ratio of 269nm & 266nm, respectively, using hexane as the reference . the acceptance criteria for the ratio is NLT 1.5 but always found that 1.3 ,why what i can do for the resolving the problem is it necessary to integrate exat at 269nm and 266nm ehat is limit of tollerence