how do start hplc method development?
how much sample required for developing the method
development?
Answers were Sorted based on User's Feedback
Answer / sushant
1.HPLC method development start with literature servey to
get rough idea abt development
2.then find out functionallity, molecular structure,
Polarity of particular API or molecule. from which we can
decide method development phase RPC or NPC.
3.Find out Solubility, Max UV absorbace so that we be
assure about respective diluent & wavelegh for method
development.
if we will not get UV absorbace then there is dervatisation
method.
4.Mostly many oragnic drug molecules are polar or mid polar
so we usually go for RPC.
5.satrt with mobile phase composition Use water & MEthanol
or ACN according to sollubillity in 50:50 ratio.
6.Select Column -start with C-18 column make HYpersil or
Inertsil(3 microne x 4.6 ID )10-15 cm
7. feed other parameter like wavelength,ambient Temp,
1ml/min Flow rate
8.About sample preparation for assay preparation use conc
in between 100-500 ppm as per peak response.
9.If there are more than 10 molecule in particular API go
for gardient method for better resolution & short run time
while developing any method always follow ICH parameter abt
Tailing factor, Capacity factor, Assymetry, thereotical
plates.Change MP composition ,Column.
Once you get better result then for sharpness or to save
time adjust flowrate & increase Temp.
10.If our desired peak isnot retain on water /ACN/Methanol
composition MP then go for Buffer. Phosphate buffer is
widly use buffer for its PKa value,easily Available, gives
more Hydrophillic effect, stable or non degradable for 2-3
days.
...............Before starting any method development on
HPLC Flush the system with water, check pressure, saturate
column with MP,check callibration status, lamp Hours.
Always keep data about whatever trials you have done during
method development.
Method Development is very challenging task so always be
logical before any changes in any parameter.
Mostly For HPLC method Development 10 gm sample is enough
but for overall development max 100 gm sample is sufficient.
Its a tentative quantity.
After method development method validation must be done.
| Is This Answer Correct ? | 237 Yes | 7 No |
Answer / kizar ahamed
Method development starts with solubility of the
compound,if itis soluble in water v can try with reverse
phase if is is not v should use normal phase
chromatography.
2)selection of wavelength
3) column selection
4)Gradient setting(Always develope the method by using
gradient than isocratic)
| Is This Answer Correct ? | 71 Yes | 25 No |
Answer / shantaram naik
depending upon response of peak prepare sample
concentration but ideally it is 10 ppm
| Is This Answer Correct ? | 1 Yes | 40 No |
What are the factors effect peak shape in HPLC ?
2 Answers College School Exams Tests,
Why Urasil and Caffeine use for HPLC wavelength accuracy calibration. which is most suitable, Why? other than these which can be used?
what is the difference b/w water content & moisture content
why sre you used Potassium hydrogen phthalate in standarisation of 1N NaOH and 0.1 N Perchloric Acid?
1 Answers Cipla, Elder, Vamsi Labs,
In the HPLC Calibration done wavelength accuracy done between 200nm-280nm .but not done remaining 300-400nm not done ?
wat is the role of michelson interferometer in FTIR and He-Ne lase.detailed answer required in simple language
In C8,C18....etc have 4&5 etc... microns Columns is there like that how to separate Chiral columns as per Micron size?
If we have 5 strength which is not dose proportinate and excipients also diffrent in each strength then how we can proceed for Force degradation? and excipient are same but not dose propotinate the how FD?
In HPLC Chromatographic separation Resolution equation is Rs = 1.18 (tR2 –tR1)/wh1+Wh2, Where, tR2 and tR1 is run time of eluent ratios of the distances from the point of application to the centers of the spots and the distance travelled by the solvent front from the point of application (retardation factor), and Wh1and Wh2 , is peak width of at half height. What is 1.18 ?
For Recent draft guidelines for Dissolution test apparatus calibration, why removing the salicylate tables?
why plane of polarized light rotated by opticaly active compunds?