A pressure transmitter with a range of 0-100 bar.The
accuracy of the instrument is 1% of upper range value?
The pressure indicating transmitter is showing the system
pressure is 50 bar.
b. When you include the error, what is the minimum possible
pressure in our fuel gas system?
i. 49.5 bar
ii. 49 bar
iii. 1 bar
iv. 0.5 bar
We want to change the filter cartridge. We must depressure
the fuel gas system.
c. After we depressure the system in preparation for
maintenance, the transmitter displays a pressure of +0.1
bar. At this time, what is the possible range of pressures
in our fuel system?
i. +0.099 to +0.101 bar
ii. 0 to +0.2 bar
iii. -0.1 to +0.3 bar
iv. -0.9 to +1.1 bar
Answers were Sorted based on User's Feedback
Answer / waseem ahmad meer
For the first question, the answer is 49 bar.
For the second question, the answer is -0.9 to +1.1 bar.
Is This Answer Correct ? | 5 Yes | 1 No |
Answer / jaydeepplc
send me practical approach answers,with example ..
Is This Answer Correct ? | 0 Yes | 0 No |
Answer / syamesh
FOR THE FIRST QUESTION ANS IS 49 BAR AND FOR THEW SECOND ANSWER IS -O.1BAR
Is This Answer Correct ? | 0 Yes | 3 No |
What is the procedure of mounting new positioner on pneumatic control valve?
How to convert gpm into inches of eater.I nerd formula with example so pls help me.
We have a Yokogawa Micro-XL DCS, which is having around 300 I/O, but whenever there is any wire gets grounded in field, the 24v supply to all loops gets dipped & all the indications at DCS goes to IOP(shows input open), at present we have fixed 500mA fuses in all the lines but which does not working. Does it required to replace with lower amps fuse, then how much or is there any other solution for this.
diff between neutral ,ground,phase , earthing and shielding?
what is beta ratio?
I want to apply for the universities in US,but i got the GRE score of (650+300=950)TOEFL 69/120.Is this score sufficient for applying universities please i am married i need to go to US because my husband didn't get H1 yet.please send me suggestion
What is controller? What is transmitter?
thru which signal does µP communicate with memory
what are the basic instructions used in simatic s7-300 plc?
5 Answers Dr Reddys, HZL, Siemens,
how can we tune a PID. If it is by zeiger nichols method..please explain briefly with help of an example.
what is control valve and how is work
Why we use 250 ohm resistor in series with HART protocol communication?
13 Answers Chemtrols, Instec, Sabic,