in hplc calibration injector linearity pass,it means inj and det is good. why to perform detector linearity is it required, if it is required,what is difference between injector and detector linearity in hplc calibration
Answers were Sorted based on User's Feedback
Answer / sudhakar tambe (gangapur)
injector linearity is perform for accuracy of injector or syringe volume accuracy. and detector linearity perform for got the same res ponce of area at specified wavelength.
Is This Answer Correct ? | 14 Yes | 0 No |
Answer / ch.murali krishna
In detector lineariry inj vol same but different conc to check the detector performance of the system
In inj linearity conc same inj vol different to check the linearity of injection 💉
Is This Answer Correct ? | 5 Yes | 0 No |
Answer / bhanumurthy.s
Both are required.In Injection volume linearity injection volume changed where as in Detector linearity constant injection volume with different Concentrations.
Is This Answer Correct ? | 4 Yes | 1 No |
Answer / dave
but Bhanumurthy.S sir
during the injector linearity performance test, as the volume change the concentration also change, it is done by system
where as in the detector linearity we manually change the concentration so ultimately we also perform detector linearity during the injector linearity test.
Is This Answer Correct ? | 0 Yes | 3 No |
Why PH range from 0to14.
2 Answers College School Exams Tests, Reddy Labs, Wings Customer Care,
why peak purity not considered in GC?
4 Answers DRL, Reddy Labs, Vimta Labs,
What is the diffeence between assay and uniformity of content
what is the difference between lod and loq?
Many times I don't got a caffeine peak in calibration of hplc using guard column ❓
how can you decide API molucule acedic or basic compound by test?
wat is reverse phase in hplc
How to select buffer for HPLC mobile phase
4 Answers Cadila, Cipla, Lupin,
what is the different in Total ash, sulphated ash, acid insoluble ash, alcoholic or non alcoholic ash?
In case of dissolution test, if there is variation within 6 individual units however the batch is complying as per S1 criteria, is it necessary to do investigation for the same? What should be the criteria for such variation?
What are usp limits for total impurities,unknown impurity
In HPLC when we use Isocratic method, Gradient method in hplc?