How to calculate LOD & LOQ in HPLC Validation Method
Answers were Sorted based on User's Feedback
Answer / abdulkalim
Based on standard deviation of the response and the slope
obtained will be calculated as:
Limit of Detection (LOD):
DL= 3(SD/Slope)
Limit of quantification(LOQ):
DL=10(SD/Slope)
Is This Answer Correct ? | 231 Yes | 20 No |
Answer / m
There are three different method indiated in ICH guidelines
for determination of LOD and LOQ.
1 Visual evaution method - For non instrument and sometimes
for instrumental also.
2 Based on residual standard deviation of the response and
the slope
obtained will be calculated as:
Limit of Detection (LOD):
DL= 3(SD/Slope)
Limit of quantification(LOQ):
DL=10(SD/Slope)
3 Signal to noise ratio method
But residual standard deviation method is more reliable
method for determination of LOD and LOQ value
Is This Answer Correct ? | 138 Yes | 17 No |
There are three different method indiated in ICH guidelines
for determination of LOD and LOQ.
1 Visual evaution method - For non instrument and sometimes
for instrumental also.
2 Slope Method:Based on residual standard deviation of the
response and
the slope
obtained will be calculated as:
Limit of Detection (LOD):
DL= 3(RESIDUAL SD/Slope)
Limit of quantification(LOQ):
DL=10(RESIDUAL SD/Slope)
3 Signal to noise ratio method
Limit of Detection (LOD):
S/N NLT 3
Limit of quantification(LOQ):
S/N NLT 10
Slope and S/N methods are more reliable
method for determination of LOD and LOQ value
Is This Answer Correct ? | 58 Yes | 11 No |
Answer / sunil.madhukar rao kulkarni
We can calculate LOD and LOQ values using linearity
graph,insted of sigma value STEYX can be considered for
more acuurate vale of LOD and LOQ
where STEYX : Standard error of the pridected Y-value for
each X in a regression
Final calulation given as below :
LOD = STEYX x 3/ Slope
LOQ = STEYX x 10 /Slope
Is This Answer Correct ? | 44 Yes | 10 No |
Answer / ashutosh kumar srivastava
Signal of noise ratio X 3 = LOD
LOD X 3.33 =LOQ
Is This Answer Correct ? | 33 Yes | 12 No |
Limit of Detection (LOD):
• Based on visual limitation
– Visual evaluation may be used for non-instrumental
methods but may also be used with instrumental methods.
– The detection limit is determined by the analysis
of samples with known concentrations of analyte and by
establishing the minimum level at which the analyte can be
reliably detected.
• Based on signal-to-noise
– This approach can only be applied to analytical
procedures which exhibit baseline noise.
– Determination of the signal-to-noise ratios is
performed by comparing measured signals from samples with
known low concentrations of analyte with those of blank
samples an establishing the minimum concentration at which
the analyte can be reliably detected. A signal-to-noise
ratio between 3 or 2:1 is generally considered acceptable
for estimating the detection limit.
• Based on the standard deviation of the response
and the slope
– The detection limit = 3.3 s / S
– Where s is the standard deviation of the response
and S is slope of the calibration curve.
– The estimate of S may be carried out in a variety
of ways, for example:
• Based on the standard deviation of the blank
• Based on the calibration curve
Limit of Quantification (LOQ):
• Based on visual limitation
– Visual evaluation may be used for non-instrumental
methods but may also be used with instrumental methods.
– The quantification limit is determined by the
analysis of samples with known concentrations of analyte
and by establishing the minimum level at which the analyte
can be quantified with acceptable accuracy and precision.
• Based on signal-to-noise
– This approach can only be applied to analytical
procedures which exhibit baseline noise.
– Determination of the signal-to-noise ratios is
performed by comparing measured signals from samples with
known low concentrations of analyte with those of blank
samples an establishing the minimum concentration at which
the analyte can be reliably quantified. A typical signal-to-
noise ratios is 10:1.
• Based on the standard deviation of the response
and the slope
– The quantification limit = 10s / S
– Where s is the standard deviation of the response
and S is slope of the calibration curve.
– The estimate of S may be carried out in a variety
of ways, for example:
• Based on the standard deviation of the blank
• Based on the calibration curve
Is This Answer Correct ? | 21 Yes | 4 No |
Answer / dr.praveen choubey
by HPLC LOD @ LOQ calculated by S/N ratio that is the best
method.LOD s/N NLT 3 and LOQ S/N NLT 10 and also calculated
by visual evalution.
Is This Answer Correct ? | 11 Yes | 2 No |
Answer / nikhil gandhi
Based on ICH Guideline there are many method given to
identify LOD and LOQ, but most useful method is slope and
residual standard deviation is used.
LOD : 3.3 X sigma/ Slope
LOQ : 10 X sigma/ Slope
Sigma : Residual Standard Deviation
Is This Answer Correct ? | 12 Yes | 4 No |
Answer / rakesh p. kotkar
Based on standard deviation of the response and the slope
obtained will be calculated as:
Limit of Detection (LOD):
DL= 3.3(ASD/Slope)
Limit of quantification(LOQ):
DL=10(ASD/Slope)
Is This Answer Correct ? | 10 Yes | 3 No |
Answer / pruthvi
why we have to take 3 for LOD and 10 for LOQ? why not other numbers?
Is This Answer Correct ? | 6 Yes | 0 No |
why digestion require in icpms?
Which parameter require to do for analytical method equivalency?
if content uniformity passing but dissolution varrying then what is next step?
how can you do the analysis of compound given to u by hplc
WHEN AN UN KNOWN SAMPLE IS GIVEN TO U,HOW WILL U START THE ANALYSIS(ie FIRST WHICH TEST TO BE DONE,NEXT,THEN NEXT),HOW WILL U DECIDE WHICH TYPE OF ANALYSIS TO BE DONE FOR IT?
Hi sir if any product monograph not given known impurities (may be 5 impurities) specifications then how we require to proceed for inhouse formulation?
As per ICH related substances stability trend limit from initial to shelf life
0 Answers Medley Pharmaceuticals, Micro Labs,
how to hplc calibrated
We use Potassium Dichromate solution for the calibration of UV-Visible Spectrophotometer in UV region. My question is to calibrate visible region which solution could be used in photometric accuracy
What is the difference between Spectroscopy and spectrophotometry,What are the applications?
4 Answers Caplin Point, Cipla,
What is difference b/w IR&NIR
why we are using buffer solutionas as a mobile phase in hplc.