Answer / s b singh
59.12 mV is equivalent to 1pH. this may check in USP also
| Is This Answer Correct ? | 58 Yes | 4 No |
Answer / pratap rasam
Ideally as per Nerst Equation 1 pH unit=59.5 mv. But we are
all the time dealing with non-ideal situations only and
thus following table should be dealt with
pH mv
1 357.0
2 297.0
3 238.0
4 178.5
5 119.0
6 59.5
7 0.00
8 -59.5
9 -119.0
10 -178.5
11 -238.0
12 -297.0
13 -357.0
14 -416.5
| Is This Answer Correct ? | 28 Yes | 5 No |
Answer / arash.babakhanian
pH MEASUREMENT IS HAPPENED USING THE DIFFRENCE BETWEEN TWO
POTETIAL VALUE CAUSED BY THE POTENTIAL RELATED TO PROTON
EXCHANGE BETWEEN INTERNAL SOLUTION AND GLass hydrated
membrane and the potential between the external solution
and glass hydrated membrane that is called bindry potential
| Is This Answer Correct ? | 10 Yes | 12 No |
What is the formula for relative diffrence for standard solution in solution stability in validation?
effect of pore size, pore volume, partical size, column length, carbon load on retention time? what is carbon load? what is the use?
why there is a need of optical activity & specific optical activity measurement by polarimeter
Why we compare unequal concentrated solutions in heavy metals test. For example comparing the test solution with 1 ppm lead standard and telling it's less than 20ppm?
1 Answers Micro Labs, Mistair,
what is batch no. of prednisole tablet.
if content uniformity passing but dissolution varrying then what is next step?
What is the difference between GSC & GLC?
What id the resolution factor
In UV calibration we use Holmium perchlorate solution made in 1.4M perchloric acid The question is that 1.4 M perchloric acid we make is it in water or glacial acetic acid?
In BP2013, Loperamid HCl monograph. Assay by titration with 0.1N sodium hydroxide using hydrocloric acid 0.01N and reading the volume added between the 2 points of inflexion. I have a question that if the diluent solvent is ethanol is certainly consumed a amount of volume of titrant, so this volume must be eliminated on the result calculation or not apart from first point which is subtracted above.
Is it required to inject 6 injections for % RSD of a system suitability in lc or gc
what is the differnce between GMP and CGMP