Range 8 Kg/cm2, alarm to be set at 2Kg/cm2, at wat pressure
reciever switch to be calibrated? justify the ans?
Answers were Sorted based on User's Feedback
Answer / shan
we are to calibrate at the pressure switch. Pressure switch
is the pressure reciever.we have to set the pressure switch
at 2Kg/cm2. If pressure reaches 2Kg/cm2,it will alarm.
Is This Answer Correct ? | 12 Yes | 1 No |
Answer / mithilesh kumar
1st pr. switch is combination of NO,NC,COMMON.OK.
for setting alarm at 2kg/cm2.
it calibrate with DEAD WEIGHT TESTER at 2kg/cm2.during
calibration NC becomes NO and NO become NC.it check with
multimeter in continuity mode.ok.
if it's done then your calibration is ok.
then you can use as per requirement.
any comment.
most welcome
Is This Answer Correct ? | 3 Yes | 2 No |
Answer / welcome4ujd
if range is 0 to 8 kg/cm2, an d alarm needed is 2 kg/cm2.
that is 25 percent of range, set the alarm at 6 psi of a 3
to 15 psi out put.
and if it is a flow transmitter of some range,say 0 to
1000 mmwc and alarm at 250 mm
then 25 % of flow in transmitter will give square root 25
devided by 10 tht is 50 percent in the control room , so
you have to set 50percent in the receiver switch that is 9
psi for a 3 to 15 psi out put.for basic flow transmitter
to find out square rooting from field to ccr
just square root the input percentage and devide by 10
to find the field input from control room signal
devide the ccr signal percentage by 10 and square it.
Is This Answer Correct ? | 2 Yes | 7 No |
what are the different types of terminal lugs used for wiring termination?
what is pneumatic control system?
diff between dead wt calibrator n pressue comparator?
What is passive connection of instrument in different types of IO module?
what is the procedure of O2 sensor(ABB) calibration?
What is meant by GAP VOLTAGE, in vibration measurement. Normally how much GAP VOLTAGE is set and state the reason.
We have ZR22G-100-S-A-E-T-T--E-A/CV/F1 Model Zirconia Oxygen analyser which is Yokogawa make. I kept the analyser in Flue gas path at Air pre heater area. My plant is 7.5MW Biomass power plant. For a particular fuel, Coconut Husk which contains Sodium and Potassium chlorides iam not getting proper reading, Analyser output always Zero. I calibrated the analyser and it is OK. For Other fuels like Wood chips, Cholam waste, DOB, Groundnutshells, Ply wood waste and all iam getting actual percentage of Oxygen. What is the main reason behind this draw back?
The wire is disconnected from the temperature sending unit (magnetic gauge). A voltmeter connected to the wire reads 0v. What does this indicate?
any body pl suggest me the best type of transmitter for level monitoring of a distillation still which consists of highly corrosive and inflammable liquid.closed vessel height=5000mm especially in petrochemical industries
Calibration procedure for water rotameters
Where is wet leg is used?
why we use 3 wire rtd direct wiring for temperature measurement instead of temperature transmitter like bearing temperature