WHAT IS A SYMPTOM OF FAULT WITH THE SPAN OF AN INSTRUMENT?
1)ZERO-POINT AND 100% ARE CORRECT BUT NOT A MID-RANGE.
2)ZERO-POINT IS CORRECT ,BUT 1005 INPUT IS NOT GIVING 100%
OUTPUT.
3)LINEARITY PRBS.
4)1005 O/p IS CORRECT BUT ZEO POINT IS WRONG.
Answer / carl_ellis
WHAT IS A SYMPTOM OF FAULT WITH THE SPAN OF AN INSTRUMENT?
1)ZERO-POINT AND 100% ARE CORRECT BUT NOT A MID-RANGE.
2)ZERO-POINT IS CORRECT ,BUT 1005 INPUT IS NOT GIVING 100%
OUTPUT. 3)LINEARITY PRBS. 4)1005 O/p IS CORRECT BUT ZEO
POINT IS WRONG.
The correct answer is 2, but number 4 also requires a span
correction.
1) The "Zero-point and 100% are correct, but not a
mid-range value" problem is a linearity problem.
2) The "Zero-point is correct, but 100% input is not giving
100% output" is a span problem.
4) The "100% output is correct but zero point is wrong" is
a zero offset AND a span problem because when the zero point
is corrected with the addition of a bias/offset, the 100%
point will shift up or down the same amount as the zero
shifted with the bias/offset correction. So the 100% point
will no longer be correct once the zero is corrected.
Hence, the zero AND the span will need to be corrected.
Carl Ellis
Measure First
Is This Answer Correct ? | 8 Yes | 2 No |
what is the different between deflection type measurement and null type measurement in RTD?
How to calibrate electronic type level switch low?
What is the difference in clamp on type and inline type ultrasonic flow transmitter.
Why do you want to work here
What is linearity of RTD over 0 to 200 degC?
what happens if we are having very very high cavity and flashing in a control valve..??? is it stops working...immediately... or eroded...slowly
HOw To Use 24V DC in Instrument?
what is the dp across a orifice plate & how much is permanent pressure across an orifice plate.
what is the input of ph sensor and meter during the calibration time and what is the output?
Which topics are covered in Process Control?
What is i/e or o/p transmitter(ma/v)?
How to adjust the zero & span of a control valve from the posiitioner(MIL)?