Answer / ramakrishna- micro labs
In RS analysis we r analysed for impurity purpose,here we take
more concentration when compare to standard,that's why
standard area comes very less when compare to sample,so inject
standard 6 replicates the % RSD is NMT 5.0 , we give 5
replicates %RSD is NMT 2.0,the system suitability criteria
than more chances to pass for 6 replicates injections
Is This Answer Correct ? | 6 Yes | 4 No |
Answer / laxman rapolu
In Related substances, we are preparing the standard in the
unknown impurity concentration level.This is the less
concentration.for consistency purpose we should
inject.Because we are quantifying impurities against diluted
standard.This is applicable to external standard method.
Is This Answer Correct ? | 3 Yes | 2 No |
What is the acceptance criteria for moisture balance when calibrated with sodium sulfate
What is the Calcultion of S/N Ratio.
What is difference between NDA & ANDA ?
Why we select scan range from higer wavelength to lower wavelength in uv visible spectroscopy and ftir spectroscopy ? .
what is the meaning of ventillation in HPLC?
3 Answers Amneal Pharmaceuticals,
What is similarity factor and its use in api strength?
What is the procedure for investigation of an OOS (Out of specification) results?
4 Answers Biocon, Dr Reddys, Reddy Labs,
In the test for control of absorbance in UV calibration why do we use only potassium dichromate and what is purpose of taking a specified amount of 57.0-63.0mg?
In Stability testing if significant change occurs then what will be the action plan?
What is the difference between low pressure gredient and high pressure gredient in hplc instrument?
what is end capping in hplc coulmn
2 Answers Lupin, Zydus Cadila,
why you are using benzoic acid for water content determination in TEA