Answer / pharmainterview
Basic concept is you need to check in acidic,neutral and
basic range. By covering all these regions you can check
anyware.
As per usp the difference between two adjacent buffers
should not more than 4.
Regards
pharmainterviewquestions.blogspot.in
Is This Answer Correct ? | 41 Yes | 0 No |
Answer / amarsingh
Because by using this all pH buffer we can cover all range
of pH (Acidic , Neutral & Basic ) So we use this range
Is This Answer Correct ? | 16 Yes | 4 No |
Answer / pankaj shrivastava
Calibration points should cover required ph adjustment of
solution. if you use only 2 to 7 ph range for ph adjustment
then you can calibrate the ph with 2.4.7 ph.
Is This Answer Correct ? | 2 Yes | 1 No |
Answer / arash babakhanian
pH is determined from this equation in the pH meter
pH = [(Eobs-K)/0.059]n
Eobs= observed potential
K= constant paramere
0.059=slope
constant parameters k and 0.059 should be calibrated. In
fact two acidic buffers and two alkalic buffers are used to
fix slope and K constant . In this case, the relation
between Eobs and pH is analyticaly change
Is This Answer Correct ? | 1 Yes | 2 No |
If we change lenght of ph eletrode wire then it will affect on ph ?
how to pepare 6.8 pH phosphate buffer
What is stray light in uv what are units for stray light
[3/30, 13:29] Manoj P Venkatpurwar: How hplc column selection according to structure? How mobile phase buffer selection on molecule structure?
In the method validation proceedure,lod & loq calculation we use the formulae for LOD=3.3*SD/slope and for LOQ=10*SD/slope,why we use 3.3 & 10 in the formulae?
how require to select dissolution media? what is discrimination?
why we use Dichlorobenzene.nitrobezine.t-butyldi sulhate.for calibration of gc Head space
Which is the highly polar and highly non polar column in HPLC?
How can we separate Urea & semicarbazide in single HPLC method ?
how to calibrate AAS ? give me the procedure?
what is the use of tap density apparatus? why we calculate it? what is the difference between the bulk density and tap density?
1 Answers Alchemist, Alembic, Apex,
From where require to take the RLD sample?