why should we do limit of stray light in UV calibration and
why should use potassium chloride
Answers were Sorted based on User's Feedback
Answer / pankaj singh
potassium chloride is neutral for uv region so transmittance
should not be less than 100%
abs = -log10 %T
when %T is 100 means 10 power 2
means abs not less than 2
Is This Answer Correct ? | 31 Yes | 10 No |
Answer / kiran pokharkar
Stray light is nothing a beam of light which does not pass
through sample.
Interference of this light is not detrible as they give
wrong results.
It is calibrated by using 1.2%w/v potassium chloride
solution in water.
Is This Answer Correct ? | 6 Yes | 10 No |
why we use vanillin in melting range calibration
if your product is soluble in 0.1n hcl and water then which you choose as media from these 2 media?
How to calibration of the uv spectroscopy and its test?
how do you established the RRF for unknown impurity
pda detector mechanism?
1 Answers DKR, Teva Pharmaceuticals,
%lod always grater than %water content for same product??
WHAT IS UNIT OF ABSORBANCE AREA IN HPLC OR UV DETECTER
when we use pda detector in hplc instead of uv detector
Why the ph limit (1_14) Is used?
why we use x terra columns at higher ph ?
why base line not starts from zero in gc?
1 Answers Aurobindo, FDC, Reddy Labs,
what r the reasons to get peak telling?