today i am doing calibration of HPLC. when i inject the
blank it was going on -ve.(bleow then zero) why?
Answers were Sorted based on User's Feedback
Answer / b.gangatharan
This is due to improper sonication of mobile phase and improper stabilisation of column.
Regards,
B.Gangatharan
| Is This Answer Correct ? | 7 Yes | 1 No |
Answer / sudheer kumar m
may be mobile phase was not stable in that particular
method. so condition the column with mobile phase and re inject.
| Is This Answer Correct ? | 2 Yes | 2 No |
Negative peak simply indicate that your mobile phase
absorbs UV light.So use only HPLC grade solvents.Even with
HPLC grade solvents you get the negative peaks you system
might not be working properly.
Regards
M.KIRAN REDDY
| Is This Answer Correct ? | 2 Yes | 7 No |
In UV we measure absorbance , how it is happend ? explain me
what is the difference between isocratic & elution technique?
If compound Pka is 5 then which buffer you will select ?
why ph electrode is storing in KCL solution?Why the pH max is 14.0?
if the flow rate is increases then any changes in retention time in hplc
how can delete the impurity in 1-(2:6 dichloro phenyl)-2-indolenone.and desolving the impurity in which solvent but does not desolve product in solvent.
why KBr is used in FTIR SPECTROSCOPY and why is it necessary to triturate KBr during FTIR analysis?
25 Answers Hikal, Lupin, Wockhardt,
water is more polar than acetonitrile.but in reverse phase chromatography peaks elute earlier when we increse Acetonitrile.what is the reason.
What is Formula of Potency calculation on Dried basis, As is basis& Anhydrous basis.
How we can identify process related and degradation impurity in single method with short period?
what is dissolution What is unit of KF? what is the significant of pH in HPlc? WHY POTASSIUM HYDROGEN PTHALATE IS USED FOR PERCHLORIC ACID? What is difference btw UV detector & RID detector?