Answer / drena

HPLC is calibrated by the following parameters:
1) Pump calibration
2) Detector Linearity
3)Injector Precision and carryover test
4) Wavelength accuracy by using caffeine and pyrene solution
5) Determine the column oven temp. and sample compartment temp
6) Gradient test
7) Detector noise

Is This Answer Correct ?    128 Yes 13 No

how to hplc calibrated ..

Answer / ajinkya p.

In HPLC there must calibrations for Pump, Detector,injector
Pump :- Should calibrated by checking flow accuracy test, Pressure holding test.
Detector :- Should calibrated by checking Linearity test, Detector wavelength accuracy test
Auto sampler :- Should calibrated by Injector Accuracy test,
Carryover test, Injector precision test, carry over test

Is This Answer Correct ?    48 Yes 2 No

how to hplc calibrated ..

Answer / kanakadurga

calibration parameters are
1.Flow accuracy
2Temperature accuracy
3Wave length accuracy by using caffeine and erbium perchlorate
4Linearity of detector and injector
5.Carry over test
6.Drift and noice

Is This Answer Correct ?    37 Yes 6 No

how to hplc calibrated ..

Answer / b.ganesh

calibration parameters are
1.Flow accuracy
2Temperature accuracy
3Wave length accuracy by using caffeine and erbium perchlorate
4Linearity of detector and injector
5.Carry over test
6.Drift and noice
7.Gradient progromme accuracy
8.injector precision

Is This Answer Correct ?    3 Yes 1 No

Post New Answer

More Analytical Chemistry Interview Questions

what is mean by covalidation

0 Answers  


Do you have experience with organic extractions or tissue digestions?

1 Answers   GVK Biosciences,


What is method validation in quality control

1 Answers   Alchem,


What is dose dumping? why require to do?

0 Answers  


why 0.1M OR 1M Perchloric acid standardisation done after 24 hours?

1 Answers   Cipla,






How to estimate potassium?

1 Answers  


If we change lenght of ph eletrode wire then it will affect on ph ?

2 Answers  


can i use hplc detector to uplc and why?

0 Answers   Mylan, SGS,


wen v have polr solvent then which type of column use in hplc

2 Answers  


if out of two set of results of moisture analysis first 1 is passing, second one is failing, is it oos ? or vice a versa first 1 is failing and second one is passing. for information mean value to be reported and in mean value the substance is ppassing, is it oos or normaal mean value to be considered?

1 Answers   JBCPL,


How is the Flame test performed?

4 Answers   Lupin,


what is the working priniciple of ph, what are parts in ph meter

1 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)