why used three wire in RTD.
Answers were Sorted based on User's Feedback
Answer / preeti
three wire connections can eliminate connection lead resistance effects from measurements
Is This Answer Correct ? | 15 Yes | 4 No |
Answer / gomsy
third wire is used as lead wire compensation i.e it is
connected in parallel,hence effective resistence decreases
and will be helpful to calculate higher range of tempreture
Is This Answer Correct ? | 13 Yes | 4 No |
Answer / santosh r apre
three wire connections can eliminate lead resistance from
mesurements.
Is This Answer Correct ? | 8 Yes | 2 No |
Answer / sam.r
Most RTD’s incorporate a third wire with resistance R3.
This configuration provides one connection to one end and
two to the other end of the sensor. Connected to an
instrument designed to accept 3-wire input, compensation is
achieved for lead resistance and temperature change in lead
resistance. This is the most commonly used configuration.
If three identical type wires are used and their lengths are
equal, then R1 = R2 = R3. By measuring the resistance
through leads 1, 2 and the resistance element, a total
system resistance is measured (R1 + R2 + RE ). If the
resistance is also measured through leads 2 and 3 (R2 + R3),
we obtain the resistance of just the lead wires, and since
all lead wire resistances are equal, subtracting this value
(R2 + R3) from the total system resistance (R1 + R2 + RE)
leaves us with just RE, and an accurate temperature
measurement has been made.
Is This Answer Correct ? | 5 Yes | 1 No |
Answer / seshu
The connecting wires long,resistance of wire changes as ambient temp fluccuates.to eleminate this problem we use 3 wire rtd
Is This Answer Correct ? | 9 Yes | 10 No |
the RTD is
installed some distance away from the transmitter. Since
the connecting
wires are long, resistance of the wires changes as ambient
temperature
fluctuates. The variations in wire resistance would
introduce an error in the
transmitter. To eliminate this problem, a three-wire RTD is
used
Is This Answer Correct ? | 2 Yes | 3 No |
How does a redundant controller comes in line when the main controller fails??
Why rtd out put is ohm how create the ohm
explain how magnetic flowmeter works
what does mean by analog output trim and where this one used for?
what effects a surging can produce in the compressor
why use MMWC unit
what r all the type of viberation measurement measuring in turbin? what's the output of viberation sensor?
what is the best method to measuring level in high pressure and temperture thank?
i have 80cm water tank.how can i use DPT as level mesurment.pls give specs for that..? We have air comp.from that moisture comeout to one tank.i want to mesure level of water in that tank and then after operat one sov to dicharge water after high level.
what is temperature compensation?
Any body have know try work calculation ???
Any one can help me to identified, how can we interchange the device in FF