How can v calibrate ftir ? Plz explain me
Answers were Sorted based on User's Feedback
Answer / a.sudhakar
FT-IR IS CALIBRATED BY USING POLYSTYRENE FILM
PUT THE POLYSTYRENE FILM INTO THE SAMPLE HOLDER
AND SCAN IT FIXED WAVE NUMBERS ARE OBTAINED THEY ARE
3060
2849
1942
1601
1583
1154
1028
ACCEPTANCE CRITERIA IS +/- 1 CM
RESOLUTION: THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 2870 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 2849.5 CM SHALL BE MORE THAN 0.33
THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 1589 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 1583 CM SHALL BE MORE THAN 0.08.
| Is This Answer Correct ? | 29 Yes | 1 No |
Answer / t n rao
FTIR IS CALIBRATED BY USING POLYSTIRENE FILM THICKNESS OF
0.04mm(AS PER USP)0.05 mm(AS PER IP. THE RESOLUTION IN
ABSORBANCE B/T 2849.5 AND 2870 IS 0.33, IN B/T 1583 AND
1587 IS 0.08.
FREEQUENCY : EVERY DAY BEFORE USING AND MAJOR MAINTENANCE
JOB.
| Is This Answer Correct ? | 8 Yes | 8 No |
How do you give training to a new chemist
What is difference b/t uv detector and rid
why we are calibrating autopolarimeter with quartz control plates
What is the meaning of 624 in DB-624 Capillary column used in GC?
what is difference between moisture contents and water contents ?
12 Answers Cipla, Hetero, USM,
Why we are using only KBr in IR spectrometer why not KCl
for which product require to do content uniformity? what is limit of cu?
Why pH range is maximum 14.0
please explian me about area normalisation method,diluted standard method,impurity vs impurity and which one has to be selected in method development
How to establish pKa for a new API molecule ?
2 Answers Neuland, Sun Pharma,
what is difference between UV - VISIBLE MODEL NO like 1600,1601,1700 etc ? plz explain me
0 Answers Graviti Pharmaceuticals,
How do you show that equal quanitities of two isomers provide the same response using a PDA