How can v calibrate ftir ? Plz explain me
Answers were Sorted based on User's Feedback
Answer / a.sudhakar
FT-IR IS CALIBRATED BY USING POLYSTYRENE FILM
PUT THE POLYSTYRENE FILM INTO THE SAMPLE HOLDER
AND SCAN IT FIXED WAVE NUMBERS ARE OBTAINED THEY ARE
3060
2849
1942
1601
1583
1154
1028
ACCEPTANCE CRITERIA IS +/- 1 CM
RESOLUTION: THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 2870 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 2849.5 CM SHALL BE MORE THAN 0.33
THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 1589 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 1583 CM SHALL BE MORE THAN 0.08.
| Is This Answer Correct ? | 29 Yes | 1 No |
Answer / t n rao
FTIR IS CALIBRATED BY USING POLYSTIRENE FILM THICKNESS OF
0.04mm(AS PER USP)0.05 mm(AS PER IP. THE RESOLUTION IN
ABSORBANCE B/T 2849.5 AND 2870 IS 0.33, IN B/T 1583 AND
1587 IS 0.08.
FREEQUENCY : EVERY DAY BEFORE USING AND MAJOR MAINTENANCE
JOB.
| Is This Answer Correct ? | 8 Yes | 8 No |
why equlirium conatant of water have 14 value
in which situation require to use paddle and basket?
hii, i'm had a sample of 1L that gone through extraction & evaporation. i used 0.5ml of methanol for reconstitution before injecting the sample to HPLC. i made a calibration curve using a standard solution but now i'm confused about how to display the concentration i got from the sample. for ex. i got 0.2ug/ml how i'm gonna explane this is for the 1L ? please help .. i'll really appreciate it thanks
How NMR spectra can be calibrated in presence or absence of TMS?
What is the formula for relative diffrence for standard solution in solution stability in validation?
some product having water content method with pyridine and ethylene glycol mixture instead of methanol? how they are selecting and methanol is not suitable solvent how ?
what is diffrence between Specific optical rotation, optical rotation?
how can you determine a compound acidic or basic?
How we will diapose UV (detector) lamp?
Why we are use 0.005M Sulphuric acid, in control of absorbance performance test UV Spectrophotometer calibration
whate is difference between calibration and validation
11 Answers Alkem Labs, Cipla, Ranbaxy, Torrent Pharma, Zydus Cadila,
how does h2so4 prevent oxidation of mohr salt before it is titrated?