Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...


How can v calibrate ftir ? Plz explain me

Answers were Sorted based on User's Feedback



How can v calibrate ftir ? Plz explain me..

Answer / a.sudhakar

FT-IR IS CALIBRATED BY USING POLYSTYRENE FILM

PUT THE POLYSTYRENE FILM INTO THE SAMPLE HOLDER
AND SCAN IT FIXED WAVE NUMBERS ARE OBTAINED THEY ARE
3060
2849
1942
1601
1583
1154
1028
ACCEPTANCE CRITERIA IS +/- 1 CM

RESOLUTION: THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 2870 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 2849.5 CM SHALL BE MORE THAN 0.33

THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 1589 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 1583 CM SHALL BE MORE THAN 0.08.

Is This Answer Correct ?    29 Yes 1 No

How can v calibrate ftir ? Plz explain me..

Answer / t n rao

FTIR IS CALIBRATED BY USING POLYSTIRENE FILM THICKNESS OF
0.04mm(AS PER USP)0.05 mm(AS PER IP. THE RESOLUTION IN
ABSORBANCE B/T 2849.5 AND 2870 IS 0.33, IN B/T 1583 AND
1587 IS 0.08.
FREEQUENCY : EVERY DAY BEFORE USING AND MAJOR MAINTENANCE
JOB.

Is This Answer Correct ?    8 Yes 8 No

Post New Answer

More Analytical Chemistry Interview Questions

How do you give training to a new chemist

3 Answers   Reddy Labs,


What is difference b/t uv detector and rid

5 Answers  


why we are calibrating autopolarimeter with quartz control plates

1 Answers  


What is the meaning of 624 in DB-624 Capillary column used in GC?

5 Answers   Aurobindo,


what is difference between moisture contents and water contents ?

12 Answers   Cipla, Hetero, USM,


Why we are using only KBr in IR spectrometer why not KCl

10 Answers   Dr Reddys,


for which product require to do content uniformity? what is limit of cu?

0 Answers  


Why pH range is maximum 14.0

6 Answers   UPL,


please explian me about area normalisation method,diluted standard method,impurity vs impurity and which one has to be selected in method development

0 Answers   Pfizer,


How to establish pKa for a new API molecule ?

2 Answers   Neuland, Sun Pharma,


what is difference between UV - VISIBLE MODEL NO like 1600,1601,1700 etc ? plz explain me

0 Answers   Graviti Pharmaceuticals,


How do you show that equal quanitities of two isomers provide the same response using a PDA

1 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)