How can v calibrate ftir ? Plz explain me
Answers were Sorted based on User's Feedback
Answer / a.sudhakar
FT-IR IS CALIBRATED BY USING POLYSTYRENE FILM
PUT THE POLYSTYRENE FILM INTO THE SAMPLE HOLDER
AND SCAN IT FIXED WAVE NUMBERS ARE OBTAINED THEY ARE
3060
2849
1942
1601
1583
1154
1028
ACCEPTANCE CRITERIA IS +/- 1 CM
RESOLUTION: THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 2870 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 2849.5 CM SHALL BE MORE THAN 0.33
THE DIFFERENCE BETWEEN THE ABSORBANCE AT THE
ABSORBATION MINIMUM AT ABOUT 1589 CM AND THE ABSORBATION
MAXIMUM AT ABOUT 1583 CM SHALL BE MORE THAN 0.08.
Is This Answer Correct ? | 29 Yes | 1 No |
Answer / t n rao
FTIR IS CALIBRATED BY USING POLYSTIRENE FILM THICKNESS OF
0.04mm(AS PER USP)0.05 mm(AS PER IP. THE RESOLUTION IN
ABSORBANCE B/T 2849.5 AND 2870 IS 0.33, IN B/T 1583 AND
1587 IS 0.08.
FREEQUENCY : EVERY DAY BEFORE USING AND MAJOR MAINTENANCE
JOB.
Is This Answer Correct ? | 8 Yes | 8 No |
What is the difference between Chromatograhic Purity And Related substances
what is the results if we use ODScolumn instead of BDS column ,(ther parameter is same for column) in assay method ?
Why we perform the ir analysis?
what is the film thickness in ftir
in dissolution why pool sample needed? in which type of drug pool sample need?
what is difference between GC & GC-HS
What is difference between" LOD" AND "VACCUME LOD"
How you collect dissolution media while in stirring condition of a basket or a paddle?
why we measure PH of a solution in PH only why not POH ?
1 Answers Stellence Pharm Science,
what is difference in ods I ods II is this is realy from sorbent / packing characteristics or from other means
why we use caffeine for hplc calibration?
[3/28, 20:52] Manoj P Venkatpurwar: how many impurities require to inject in assay specificity that how we can find out? and in Rs also how?